Article Type
Changed
Wed, 12/14/2016 - 10:29
Display Headline
Assay Could Speed Diagnosis of Bacterial Sepsis

A testing method that uses positive blood cultures and an automated DNA-based polymerase chain reaction and microarray system accurately identified bacteria in sepsis much faster than a standard culture-based process.

The test identified bacterial species in patients with suspected sepsis with 95% sensitivity and 99% specificity, with a mean turnaround time of 23 hours—compared with 41.5–48 hours for the standard culture-based method, according to a report.

Although the sepsis assay is a “major advance” that encompasses the best of nucleic acid and standard culture-based methods, it is unknown whether determining the species of a pathogen 18 hours earlier than usual will “translate into demonstrable clinical benefit commensurate to the cost of undertaking the additional test,” commented Dr. Shin Lin of Stanford (Calif.) University and Dr. Samuel Yang of Johns Hopkins University, Baltimore, who were not involved in the study (Lancet 2009 Dec. 10 [doi:10.1016/S0140-6736(09)61791-8]).

In the study, Dr. Päivi Tissari of the Helsinki University Hospital Laboratory and colleagues tested the Prove-it sepsis assay, manufactured by Helsinki-based Mobidiag, against standard blood culture and pathogen identification. The analysis included 3,318 blood samples from patients with suspected sepsis at two large academic medical centers (Lancet 2009 Dec. 10 [doi:10.1016/S0140-6736(09)61569-5]).

The assay identifies more than 50 species of gram-positive and gram-negative bacteria that cause most cases of sepsis. A total of 2,107 blood culture samples tested positive, including 1,807 that were covered by the sepsis assay. The researchers compared DNA sequences of topoisomerase and 16S rRNA genes and original microbiologic laboratory data for samples when the results differed between the assay and standard blood culture method.

For organisms that could be detected with the sepsis assay, the results of the assay were between 93% and 100% concordant with the results of blood culturing for all species except one. The assay identified 133 of the 163 coagulase-negative staphylococci that were identified through blood culture.

False-positive results were identified in 52 of the 3,318 samples put through the assay. Those 52 false positives included 34 that were due to contamination or software failure, 11 with more bacterial species detected than with conventional blood culture, 3 with Staphylococcus epidermidis reported, 3 attributed to cross-hybridization between species, and 1 sample in which the assay also detected Bacteroides fragilis.

False-negative results occurred in 34 samples due to inadequate sensitivity for certain species, and in 60 samples because the sepsis assay did not detect all the bacteria it should have. The assay also had difficulty in resolving species in polymicrobial samples.

The median difference in turnaround time between the Prove-it assay and the reference method for 39 samples was 18 hours 19 minutes.

The assay provided 100% sensitivity and specificity for methicillin-resistant Staphylococcus aureus, although it is the only type of antibiotic resistance testing that can be performed with the assay.

“The Prove-it sepsis assay cannot replace standard methods but could have a role alongside them,” Dr. Lin and Dr. Yang wrote.

Five of the investigators are employees of Mobidiag, which provided the equipment and reagents for the assay. None of the other researchers had conflicts of interest. Dr. Lin and Dr. Yang had no conflicts of interest to report. The study was performed without outside funding.

My Take

Sepsis Assay Requires More Study

As Dr. Lin and Dr. Yang note, until someone shows that being able to diagnose bacterial sepsis 18 hours faster makes a difference in patient care, it's not clear that this assay would be a clinically important advance.

One could speculate about other possible benefits, such as reductions in empiric antibiotic use, antibiotic exposure and resistance, or costs for medication and lab testing. But someone needs to demonstrate such benefits. Otherwise, who cares?

FRANK MICHOTA, M.D., is the Director of Academic Affairs in the Department of Hospital Medicine at the Cleveland Clinic. He reports no relevant conflicts of interest.

Article PDF
Author and Disclosure Information

Publications
Topics
Author and Disclosure Information

Author and Disclosure Information

Article PDF
Article PDF

A testing method that uses positive blood cultures and an automated DNA-based polymerase chain reaction and microarray system accurately identified bacteria in sepsis much faster than a standard culture-based process.

The test identified bacterial species in patients with suspected sepsis with 95% sensitivity and 99% specificity, with a mean turnaround time of 23 hours—compared with 41.5–48 hours for the standard culture-based method, according to a report.

Although the sepsis assay is a “major advance” that encompasses the best of nucleic acid and standard culture-based methods, it is unknown whether determining the species of a pathogen 18 hours earlier than usual will “translate into demonstrable clinical benefit commensurate to the cost of undertaking the additional test,” commented Dr. Shin Lin of Stanford (Calif.) University and Dr. Samuel Yang of Johns Hopkins University, Baltimore, who were not involved in the study (Lancet 2009 Dec. 10 [doi:10.1016/S0140-6736(09)61791-8]).

In the study, Dr. Päivi Tissari of the Helsinki University Hospital Laboratory and colleagues tested the Prove-it sepsis assay, manufactured by Helsinki-based Mobidiag, against standard blood culture and pathogen identification. The analysis included 3,318 blood samples from patients with suspected sepsis at two large academic medical centers (Lancet 2009 Dec. 10 [doi:10.1016/S0140-6736(09)61569-5]).

The assay identifies more than 50 species of gram-positive and gram-negative bacteria that cause most cases of sepsis. A total of 2,107 blood culture samples tested positive, including 1,807 that were covered by the sepsis assay. The researchers compared DNA sequences of topoisomerase and 16S rRNA genes and original microbiologic laboratory data for samples when the results differed between the assay and standard blood culture method.

For organisms that could be detected with the sepsis assay, the results of the assay were between 93% and 100% concordant with the results of blood culturing for all species except one. The assay identified 133 of the 163 coagulase-negative staphylococci that were identified through blood culture.

False-positive results were identified in 52 of the 3,318 samples put through the assay. Those 52 false positives included 34 that were due to contamination or software failure, 11 with more bacterial species detected than with conventional blood culture, 3 with Staphylococcus epidermidis reported, 3 attributed to cross-hybridization between species, and 1 sample in which the assay also detected Bacteroides fragilis.

False-negative results occurred in 34 samples due to inadequate sensitivity for certain species, and in 60 samples because the sepsis assay did not detect all the bacteria it should have. The assay also had difficulty in resolving species in polymicrobial samples.

The median difference in turnaround time between the Prove-it assay and the reference method for 39 samples was 18 hours 19 minutes.

The assay provided 100% sensitivity and specificity for methicillin-resistant Staphylococcus aureus, although it is the only type of antibiotic resistance testing that can be performed with the assay.

“The Prove-it sepsis assay cannot replace standard methods but could have a role alongside them,” Dr. Lin and Dr. Yang wrote.

Five of the investigators are employees of Mobidiag, which provided the equipment and reagents for the assay. None of the other researchers had conflicts of interest. Dr. Lin and Dr. Yang had no conflicts of interest to report. The study was performed without outside funding.

My Take

Sepsis Assay Requires More Study

As Dr. Lin and Dr. Yang note, until someone shows that being able to diagnose bacterial sepsis 18 hours faster makes a difference in patient care, it's not clear that this assay would be a clinically important advance.

One could speculate about other possible benefits, such as reductions in empiric antibiotic use, antibiotic exposure and resistance, or costs for medication and lab testing. But someone needs to demonstrate such benefits. Otherwise, who cares?

FRANK MICHOTA, M.D., is the Director of Academic Affairs in the Department of Hospital Medicine at the Cleveland Clinic. He reports no relevant conflicts of interest.

A testing method that uses positive blood cultures and an automated DNA-based polymerase chain reaction and microarray system accurately identified bacteria in sepsis much faster than a standard culture-based process.

The test identified bacterial species in patients with suspected sepsis with 95% sensitivity and 99% specificity, with a mean turnaround time of 23 hours—compared with 41.5–48 hours for the standard culture-based method, according to a report.

Although the sepsis assay is a “major advance” that encompasses the best of nucleic acid and standard culture-based methods, it is unknown whether determining the species of a pathogen 18 hours earlier than usual will “translate into demonstrable clinical benefit commensurate to the cost of undertaking the additional test,” commented Dr. Shin Lin of Stanford (Calif.) University and Dr. Samuel Yang of Johns Hopkins University, Baltimore, who were not involved in the study (Lancet 2009 Dec. 10 [doi:10.1016/S0140-6736(09)61791-8]).

In the study, Dr. Päivi Tissari of the Helsinki University Hospital Laboratory and colleagues tested the Prove-it sepsis assay, manufactured by Helsinki-based Mobidiag, against standard blood culture and pathogen identification. The analysis included 3,318 blood samples from patients with suspected sepsis at two large academic medical centers (Lancet 2009 Dec. 10 [doi:10.1016/S0140-6736(09)61569-5]).

The assay identifies more than 50 species of gram-positive and gram-negative bacteria that cause most cases of sepsis. A total of 2,107 blood culture samples tested positive, including 1,807 that were covered by the sepsis assay. The researchers compared DNA sequences of topoisomerase and 16S rRNA genes and original microbiologic laboratory data for samples when the results differed between the assay and standard blood culture method.

For organisms that could be detected with the sepsis assay, the results of the assay were between 93% and 100% concordant with the results of blood culturing for all species except one. The assay identified 133 of the 163 coagulase-negative staphylococci that were identified through blood culture.

False-positive results were identified in 52 of the 3,318 samples put through the assay. Those 52 false positives included 34 that were due to contamination or software failure, 11 with more bacterial species detected than with conventional blood culture, 3 with Staphylococcus epidermidis reported, 3 attributed to cross-hybridization between species, and 1 sample in which the assay also detected Bacteroides fragilis.

False-negative results occurred in 34 samples due to inadequate sensitivity for certain species, and in 60 samples because the sepsis assay did not detect all the bacteria it should have. The assay also had difficulty in resolving species in polymicrobial samples.

The median difference in turnaround time between the Prove-it assay and the reference method for 39 samples was 18 hours 19 minutes.

The assay provided 100% sensitivity and specificity for methicillin-resistant Staphylococcus aureus, although it is the only type of antibiotic resistance testing that can be performed with the assay.

“The Prove-it sepsis assay cannot replace standard methods but could have a role alongside them,” Dr. Lin and Dr. Yang wrote.

Five of the investigators are employees of Mobidiag, which provided the equipment and reagents for the assay. None of the other researchers had conflicts of interest. Dr. Lin and Dr. Yang had no conflicts of interest to report. The study was performed without outside funding.

My Take

Sepsis Assay Requires More Study

As Dr. Lin and Dr. Yang note, until someone shows that being able to diagnose bacterial sepsis 18 hours faster makes a difference in patient care, it's not clear that this assay would be a clinically important advance.

One could speculate about other possible benefits, such as reductions in empiric antibiotic use, antibiotic exposure and resistance, or costs for medication and lab testing. But someone needs to demonstrate such benefits. Otherwise, who cares?

FRANK MICHOTA, M.D., is the Director of Academic Affairs in the Department of Hospital Medicine at the Cleveland Clinic. He reports no relevant conflicts of interest.

Publications
Publications
Topics
Article Type
Display Headline
Assay Could Speed Diagnosis of Bacterial Sepsis
Display Headline
Assay Could Speed Diagnosis of Bacterial Sepsis
Article Source

PURLs Copyright

Inside the Article

Article PDF Media