Widespread, routine use of HFOV now questioned
Article Type
Changed
Wed, 01/02/2019 - 08:22
Display Headline
High-frequency oscillatory ventilation may worsen ARDS

High-frequency oscillatory ventilation doesn’t improve and may actually worsen moderate to severe acute respiratory distress syndrome in adults, compared with standard ventilation, according to two reports published online Jan. 23 in the New England Journal of Medicine.

In two large, separate, randomized controlled trials comparing the two ventilation strategies, 1-month mortality in critically ill adults with ARDS who received high-frequency oscillatory ventilation (HFOV) was either higher or not significantly different from that in patients who received standard low tidal volume and high positive end–expiratory pressure ventilation.

Dr. Niall D. Ferguson    

In one study, which was terminated early because of the large discrepancy in short-term mortality, HFOV also was associated with higher mean airway pressures and significantly greater need for sedatives, neuromuscular blockers, and vasoactive drugs.

Both trials call into question the current widespread use of HFOV early in the course of ARDS when patients don’t show an adequate response to conventional mechanical ventilation, the two research groups noted.

HFOV, which delivers very small tidal volumes at very high rates, is thought to minimize the lung damage done by ventilation’s repeated forced opening and collapsing of lung structures. Many clinicians now use it earlier in the course of ARDS, even though there are other approaches for improving oxygenation, based solely on the results of animal studies and small trials that used outdated ventilation methods as a control. The commercial availability of HFOV equipment has accelerated this trend.

In the absence of good evidence of HFOV’s effectiveness, experts in Canada and the United Kingdom called for rigorous randomized controlled trials.

Dr. Niall D. Ferguson and his associates in the Oscillation for Acute Respiratory Distress Syndrome Treated Early (OSCILLATE) trial compared the two strategies in patients aged 16-85 years who had moderate to severe ARDS and were treated at 39 intensive care units in Canada, Saudi Arabia, the United States, Chile, and India.

The OSCILLATE steering committee terminated the trial early, after only 548 subjects had been randomized, because three consecutive interim analyses showed that 1-month mortality was consistently higher with HFOV. "The effect size was sufficiently large that we concluded that even if early HFOV did not increase mortality, it would be very unlikely to decrease mortality," wrote Dr. Ferguson of the division of critical care medicine and the departments of medicine and physiology, University of Toronto, and his colleagues.

At that time, 1-month mortality was 47% for HFOV (129 of 275 patients), compared with 35% (96 of 273 patients) in the control group (N. Engl. J. Med. 2013 Jan. 23 [doi:10.1056/NEJMoa1215554]).

This result remained robust in three further analyses of the data that controlled for numerous variables. It also persisted in several subgroup analyses, regardless of the subjects’ baseline severity of hypoxemia, the subjects’ BMI, the use or nonuse of vasopressors, or the level of experience with ventilation at each medical center.

In addition, the use of vasopressors, which had been equivalent between the two study groups at baseline at approximately 63%, increased in the HFOV group to 73% within 4 hours of beginning the procedure and to 78% the next day. In contrast, vasopressors were used in only 62% and 58%, respectively, in the control group.

The use of neuromuscular blockers followed the same pattern, increasing in the HFOV group from 27% at baseline to 46% at 4 hours and 46% at 24 hours but remaining more steady in the control group at 29%, 31%, and 26%, respectively, they reported.

Cumulatively, vasoactive agents were used in 91% and neuromuscular blockers in 83% of the HFOV group, compared with 84% and 68%, respectively, in the control group. Vasoactive agents were required for an average of 2 days longer in the HFOV group, and neuromuscular blockers were required for an average of 1 day longer, compared with the control group.

Sedatives and opioids (primarily midazolam and fentanyl) were given for the same duration to the two study groups, but doses were higher in the HFOV group.

"Our results raise serious concerns about the early use of HFOV for the management of ARDS in adults. The results of this study increase the uncertainty about possible benefits of HFOV even when applied in patients with life-threatening refractory hypoxemia," Dr. Ferguson and his associates said.

In the Oscillation in ARDS (OSCAR) study, commissioned by the U.K.’s National Institute for Health Research, two ventilation techniques were compared in 795 patients in general ICUs at 12 university hospitals, 4 university-affiliated hospitals, and 13 district general hospitals in England, Wales, and Scotland.

As this was a pragmatic study meant to reflect real-world practice, HFOV was compared against whatever conventional ventilation was used in local practice, and the ICUs were "encouraged" to use the recommended combinations of positive end-expiratory pressure and fraction of inspired oxygen values. Consequently, management of the control group varied greatly from one ICU to the next, according to the researchers, led by Dr. Duncan Young of John Radcliffe Hospital and the University of Oxford, both in Oxford.

 

 

The primary outcome, 30-day all-cause mortality, occurred in 41.7% of the HFOV group (166 of 398 patients) and in 41.1% of the control group (163 of 397 patients), a nonsignificant difference. These rates remained largely unchanged in further analyses that adjusted for several variables, Dr. Young and his colleagues said (N. Engl. J. Med. 2013 Jan. 23 [doi:10.1056/NEJMoa1215716]).

The duration of ICU stay was 17.6 days with HFOV and 16.1 days with conventional ventilation, and the duration of hospital stay was 33.1 days and 33.9 days, respectively.

"Since data collection is ongoing, we cannot yet report the longer-term outcomes (including survival and health-related quality of life)," they added.

In conclusion, "we were unable to find any benefit or harm from the use of HFOV in adult patients with ARDS. We recommend that this mode of ventilation not be used for routine care," Dr. Young and his associates said.

The OSCILLATE study was supported by the Canadian Institutes of Health Research, the King Abdullah International Medical Research Center, and Fonds de Recherche de Quebec. Dr. Ferguson’s associates reported ties to numerous industry sources. The OSCAR study was supported by the National Institute for Health Research Health Technology Assessment Programme. Dr. Young and his associates reported no relevant financial conflicts of interest.

CareFusion provided the SensorMedics HFO ventilator and technical support for the OSCILLATE trial but had no role in study design, data collection or analysis, or manuscript preparation. Inspiration Healthcare supplied the ventilators for the OSCAR study but had no role in the study design, data acquisition or analysis, or manuscript preparation.

Body

Both the OSCILLATE and the OSCAR trial raise concerns about the widespread and routine use of HFOV, said Dr. Atul Malhotra and Dr. Jeffrey M. Drazen.

    



Dr. Jeffrey M. Drazen

"However, one could argue that it is not HFOV itself but the HFOV protocols studied in these trials that were ineffective, and perhaps worse, than usual care," they wrote. Questions remain concerning this method’s usefulness after more aggressive volume resuscitation, less aggressive use of sedation and paralysis, and selection of patients who have homgeneous, recruitable lung.

"Perhaps patients with ARDS will require individualized therapy that takes into consideration their body habitus, the cause of their disease, and the mechanisms leading to lung injury ... but for now clinicians should be cautious about applying HFOV routinely in patients with ARDS," they said.

Dr. Malhotra and Dr. Drazen are with the divisions of sleep medicine and pulmonary and critical care medicine at Brigham and Women’s Hospital, Boston. Dr. Malhotra reported ties to Philips Respironics, Pfizer, Apnex, and Apnicure. These remarks were taken from their editorial accompanying the two reports (N. Engl. J. Med. 2013 Jan. 23 [doi:10.1056/NEJMe1300103]).

Author and Disclosure Information

Publications
Topics
Legacy Keywords
high-frequency oscillatory ventilation, New England Journal of Medicine, ARDS, acute respiratory distress syndrome
Author and Disclosure Information

Author and Disclosure Information

Body

Both the OSCILLATE and the OSCAR trial raise concerns about the widespread and routine use of HFOV, said Dr. Atul Malhotra and Dr. Jeffrey M. Drazen.

    



Dr. Jeffrey M. Drazen

"However, one could argue that it is not HFOV itself but the HFOV protocols studied in these trials that were ineffective, and perhaps worse, than usual care," they wrote. Questions remain concerning this method’s usefulness after more aggressive volume resuscitation, less aggressive use of sedation and paralysis, and selection of patients who have homgeneous, recruitable lung.

"Perhaps patients with ARDS will require individualized therapy that takes into consideration their body habitus, the cause of their disease, and the mechanisms leading to lung injury ... but for now clinicians should be cautious about applying HFOV routinely in patients with ARDS," they said.

Dr. Malhotra and Dr. Drazen are with the divisions of sleep medicine and pulmonary and critical care medicine at Brigham and Women’s Hospital, Boston. Dr. Malhotra reported ties to Philips Respironics, Pfizer, Apnex, and Apnicure. These remarks were taken from their editorial accompanying the two reports (N. Engl. J. Med. 2013 Jan. 23 [doi:10.1056/NEJMe1300103]).

Body

Both the OSCILLATE and the OSCAR trial raise concerns about the widespread and routine use of HFOV, said Dr. Atul Malhotra and Dr. Jeffrey M. Drazen.

    



Dr. Jeffrey M. Drazen

"However, one could argue that it is not HFOV itself but the HFOV protocols studied in these trials that were ineffective, and perhaps worse, than usual care," they wrote. Questions remain concerning this method’s usefulness after more aggressive volume resuscitation, less aggressive use of sedation and paralysis, and selection of patients who have homgeneous, recruitable lung.

"Perhaps patients with ARDS will require individualized therapy that takes into consideration their body habitus, the cause of their disease, and the mechanisms leading to lung injury ... but for now clinicians should be cautious about applying HFOV routinely in patients with ARDS," they said.

Dr. Malhotra and Dr. Drazen are with the divisions of sleep medicine and pulmonary and critical care medicine at Brigham and Women’s Hospital, Boston. Dr. Malhotra reported ties to Philips Respironics, Pfizer, Apnex, and Apnicure. These remarks were taken from their editorial accompanying the two reports (N. Engl. J. Med. 2013 Jan. 23 [doi:10.1056/NEJMe1300103]).

Title
Widespread, routine use of HFOV now questioned
Widespread, routine use of HFOV now questioned

High-frequency oscillatory ventilation doesn’t improve and may actually worsen moderate to severe acute respiratory distress syndrome in adults, compared with standard ventilation, according to two reports published online Jan. 23 in the New England Journal of Medicine.

In two large, separate, randomized controlled trials comparing the two ventilation strategies, 1-month mortality in critically ill adults with ARDS who received high-frequency oscillatory ventilation (HFOV) was either higher or not significantly different from that in patients who received standard low tidal volume and high positive end–expiratory pressure ventilation.

Dr. Niall D. Ferguson    

In one study, which was terminated early because of the large discrepancy in short-term mortality, HFOV also was associated with higher mean airway pressures and significantly greater need for sedatives, neuromuscular blockers, and vasoactive drugs.

Both trials call into question the current widespread use of HFOV early in the course of ARDS when patients don’t show an adequate response to conventional mechanical ventilation, the two research groups noted.

HFOV, which delivers very small tidal volumes at very high rates, is thought to minimize the lung damage done by ventilation’s repeated forced opening and collapsing of lung structures. Many clinicians now use it earlier in the course of ARDS, even though there are other approaches for improving oxygenation, based solely on the results of animal studies and small trials that used outdated ventilation methods as a control. The commercial availability of HFOV equipment has accelerated this trend.

In the absence of good evidence of HFOV’s effectiveness, experts in Canada and the United Kingdom called for rigorous randomized controlled trials.

Dr. Niall D. Ferguson and his associates in the Oscillation for Acute Respiratory Distress Syndrome Treated Early (OSCILLATE) trial compared the two strategies in patients aged 16-85 years who had moderate to severe ARDS and were treated at 39 intensive care units in Canada, Saudi Arabia, the United States, Chile, and India.

The OSCILLATE steering committee terminated the trial early, after only 548 subjects had been randomized, because three consecutive interim analyses showed that 1-month mortality was consistently higher with HFOV. "The effect size was sufficiently large that we concluded that even if early HFOV did not increase mortality, it would be very unlikely to decrease mortality," wrote Dr. Ferguson of the division of critical care medicine and the departments of medicine and physiology, University of Toronto, and his colleagues.

At that time, 1-month mortality was 47% for HFOV (129 of 275 patients), compared with 35% (96 of 273 patients) in the control group (N. Engl. J. Med. 2013 Jan. 23 [doi:10.1056/NEJMoa1215554]).

This result remained robust in three further analyses of the data that controlled for numerous variables. It also persisted in several subgroup analyses, regardless of the subjects’ baseline severity of hypoxemia, the subjects’ BMI, the use or nonuse of vasopressors, or the level of experience with ventilation at each medical center.

In addition, the use of vasopressors, which had been equivalent between the two study groups at baseline at approximately 63%, increased in the HFOV group to 73% within 4 hours of beginning the procedure and to 78% the next day. In contrast, vasopressors were used in only 62% and 58%, respectively, in the control group.

The use of neuromuscular blockers followed the same pattern, increasing in the HFOV group from 27% at baseline to 46% at 4 hours and 46% at 24 hours but remaining more steady in the control group at 29%, 31%, and 26%, respectively, they reported.

Cumulatively, vasoactive agents were used in 91% and neuromuscular blockers in 83% of the HFOV group, compared with 84% and 68%, respectively, in the control group. Vasoactive agents were required for an average of 2 days longer in the HFOV group, and neuromuscular blockers were required for an average of 1 day longer, compared with the control group.

Sedatives and opioids (primarily midazolam and fentanyl) were given for the same duration to the two study groups, but doses were higher in the HFOV group.

"Our results raise serious concerns about the early use of HFOV for the management of ARDS in adults. The results of this study increase the uncertainty about possible benefits of HFOV even when applied in patients with life-threatening refractory hypoxemia," Dr. Ferguson and his associates said.

In the Oscillation in ARDS (OSCAR) study, commissioned by the U.K.’s National Institute for Health Research, two ventilation techniques were compared in 795 patients in general ICUs at 12 university hospitals, 4 university-affiliated hospitals, and 13 district general hospitals in England, Wales, and Scotland.

As this was a pragmatic study meant to reflect real-world practice, HFOV was compared against whatever conventional ventilation was used in local practice, and the ICUs were "encouraged" to use the recommended combinations of positive end-expiratory pressure and fraction of inspired oxygen values. Consequently, management of the control group varied greatly from one ICU to the next, according to the researchers, led by Dr. Duncan Young of John Radcliffe Hospital and the University of Oxford, both in Oxford.

 

 

The primary outcome, 30-day all-cause mortality, occurred in 41.7% of the HFOV group (166 of 398 patients) and in 41.1% of the control group (163 of 397 patients), a nonsignificant difference. These rates remained largely unchanged in further analyses that adjusted for several variables, Dr. Young and his colleagues said (N. Engl. J. Med. 2013 Jan. 23 [doi:10.1056/NEJMoa1215716]).

The duration of ICU stay was 17.6 days with HFOV and 16.1 days with conventional ventilation, and the duration of hospital stay was 33.1 days and 33.9 days, respectively.

"Since data collection is ongoing, we cannot yet report the longer-term outcomes (including survival and health-related quality of life)," they added.

In conclusion, "we were unable to find any benefit or harm from the use of HFOV in adult patients with ARDS. We recommend that this mode of ventilation not be used for routine care," Dr. Young and his associates said.

The OSCILLATE study was supported by the Canadian Institutes of Health Research, the King Abdullah International Medical Research Center, and Fonds de Recherche de Quebec. Dr. Ferguson’s associates reported ties to numerous industry sources. The OSCAR study was supported by the National Institute for Health Research Health Technology Assessment Programme. Dr. Young and his associates reported no relevant financial conflicts of interest.

CareFusion provided the SensorMedics HFO ventilator and technical support for the OSCILLATE trial but had no role in study design, data collection or analysis, or manuscript preparation. Inspiration Healthcare supplied the ventilators for the OSCAR study but had no role in the study design, data acquisition or analysis, or manuscript preparation.

High-frequency oscillatory ventilation doesn’t improve and may actually worsen moderate to severe acute respiratory distress syndrome in adults, compared with standard ventilation, according to two reports published online Jan. 23 in the New England Journal of Medicine.

In two large, separate, randomized controlled trials comparing the two ventilation strategies, 1-month mortality in critically ill adults with ARDS who received high-frequency oscillatory ventilation (HFOV) was either higher or not significantly different from that in patients who received standard low tidal volume and high positive end–expiratory pressure ventilation.

Dr. Niall D. Ferguson    

In one study, which was terminated early because of the large discrepancy in short-term mortality, HFOV also was associated with higher mean airway pressures and significantly greater need for sedatives, neuromuscular blockers, and vasoactive drugs.

Both trials call into question the current widespread use of HFOV early in the course of ARDS when patients don’t show an adequate response to conventional mechanical ventilation, the two research groups noted.

HFOV, which delivers very small tidal volumes at very high rates, is thought to minimize the lung damage done by ventilation’s repeated forced opening and collapsing of lung structures. Many clinicians now use it earlier in the course of ARDS, even though there are other approaches for improving oxygenation, based solely on the results of animal studies and small trials that used outdated ventilation methods as a control. The commercial availability of HFOV equipment has accelerated this trend.

In the absence of good evidence of HFOV’s effectiveness, experts in Canada and the United Kingdom called for rigorous randomized controlled trials.

Dr. Niall D. Ferguson and his associates in the Oscillation for Acute Respiratory Distress Syndrome Treated Early (OSCILLATE) trial compared the two strategies in patients aged 16-85 years who had moderate to severe ARDS and were treated at 39 intensive care units in Canada, Saudi Arabia, the United States, Chile, and India.

The OSCILLATE steering committee terminated the trial early, after only 548 subjects had been randomized, because three consecutive interim analyses showed that 1-month mortality was consistently higher with HFOV. "The effect size was sufficiently large that we concluded that even if early HFOV did not increase mortality, it would be very unlikely to decrease mortality," wrote Dr. Ferguson of the division of critical care medicine and the departments of medicine and physiology, University of Toronto, and his colleagues.

At that time, 1-month mortality was 47% for HFOV (129 of 275 patients), compared with 35% (96 of 273 patients) in the control group (N. Engl. J. Med. 2013 Jan. 23 [doi:10.1056/NEJMoa1215554]).

This result remained robust in three further analyses of the data that controlled for numerous variables. It also persisted in several subgroup analyses, regardless of the subjects’ baseline severity of hypoxemia, the subjects’ BMI, the use or nonuse of vasopressors, or the level of experience with ventilation at each medical center.

In addition, the use of vasopressors, which had been equivalent between the two study groups at baseline at approximately 63%, increased in the HFOV group to 73% within 4 hours of beginning the procedure and to 78% the next day. In contrast, vasopressors were used in only 62% and 58%, respectively, in the control group.

The use of neuromuscular blockers followed the same pattern, increasing in the HFOV group from 27% at baseline to 46% at 4 hours and 46% at 24 hours but remaining more steady in the control group at 29%, 31%, and 26%, respectively, they reported.

Cumulatively, vasoactive agents were used in 91% and neuromuscular blockers in 83% of the HFOV group, compared with 84% and 68%, respectively, in the control group. Vasoactive agents were required for an average of 2 days longer in the HFOV group, and neuromuscular blockers were required for an average of 1 day longer, compared with the control group.

Sedatives and opioids (primarily midazolam and fentanyl) were given for the same duration to the two study groups, but doses were higher in the HFOV group.

"Our results raise serious concerns about the early use of HFOV for the management of ARDS in adults. The results of this study increase the uncertainty about possible benefits of HFOV even when applied in patients with life-threatening refractory hypoxemia," Dr. Ferguson and his associates said.

In the Oscillation in ARDS (OSCAR) study, commissioned by the U.K.’s National Institute for Health Research, two ventilation techniques were compared in 795 patients in general ICUs at 12 university hospitals, 4 university-affiliated hospitals, and 13 district general hospitals in England, Wales, and Scotland.

As this was a pragmatic study meant to reflect real-world practice, HFOV was compared against whatever conventional ventilation was used in local practice, and the ICUs were "encouraged" to use the recommended combinations of positive end-expiratory pressure and fraction of inspired oxygen values. Consequently, management of the control group varied greatly from one ICU to the next, according to the researchers, led by Dr. Duncan Young of John Radcliffe Hospital and the University of Oxford, both in Oxford.

 

 

The primary outcome, 30-day all-cause mortality, occurred in 41.7% of the HFOV group (166 of 398 patients) and in 41.1% of the control group (163 of 397 patients), a nonsignificant difference. These rates remained largely unchanged in further analyses that adjusted for several variables, Dr. Young and his colleagues said (N. Engl. J. Med. 2013 Jan. 23 [doi:10.1056/NEJMoa1215716]).

The duration of ICU stay was 17.6 days with HFOV and 16.1 days with conventional ventilation, and the duration of hospital stay was 33.1 days and 33.9 days, respectively.

"Since data collection is ongoing, we cannot yet report the longer-term outcomes (including survival and health-related quality of life)," they added.

In conclusion, "we were unable to find any benefit or harm from the use of HFOV in adult patients with ARDS. We recommend that this mode of ventilation not be used for routine care," Dr. Young and his associates said.

The OSCILLATE study was supported by the Canadian Institutes of Health Research, the King Abdullah International Medical Research Center, and Fonds de Recherche de Quebec. Dr. Ferguson’s associates reported ties to numerous industry sources. The OSCAR study was supported by the National Institute for Health Research Health Technology Assessment Programme. Dr. Young and his associates reported no relevant financial conflicts of interest.

CareFusion provided the SensorMedics HFO ventilator and technical support for the OSCILLATE trial but had no role in study design, data collection or analysis, or manuscript preparation. Inspiration Healthcare supplied the ventilators for the OSCAR study but had no role in the study design, data acquisition or analysis, or manuscript preparation.

Publications
Publications
Topics
Article Type
Display Headline
High-frequency oscillatory ventilation may worsen ARDS
Display Headline
High-frequency oscillatory ventilation may worsen ARDS
Legacy Keywords
high-frequency oscillatory ventilation, New England Journal of Medicine, ARDS, acute respiratory distress syndrome
Legacy Keywords
high-frequency oscillatory ventilation, New England Journal of Medicine, ARDS, acute respiratory distress syndrome
Article Source

FROM THE NEW ENGLAND JOURNAL OF MEDICINE

PURLs Copyright

Inside the Article

Vitals

Major Finding: 1-month mortality was significantly higher with HFOV (47%) than with control ventilation (35%) in the OSCILLATE study; no significant difference was seen between groups in the OSCAR study (41.7% vs 41.1%).

Data Source: Two randomized controlled trials comparing HFOV against either low tidal volume, high positive end–expiratory pressure ventilation or conventional ventilation among 548 ARDS patients in Canada, Saudi Arabia, the United States, Chile, and India and among 795 ARDS patients in England, Wales, and Scotland.

Disclosures: The OSCILLATE study was supported by the Canadian Institutes of Health Research, the King Abdullah International Medical Research Center, and Fonds de Recherche de Quebec. Dr. Ferguson's associates reported ties to numerous industry sources. The OSCAR study was supported by the National Institute for Health Research Health Technology Assessment Programme. Dr. Young and his associates reported no relevant financial conflicts of interest.