Article Type
Changed

Health care facilities and providers should complete the transition to fully disposable duodenoscopes and those with disposable components, the U.S. Food and Drug Administration announced this week after an analysis of postmarket surveillance studies was completed.

The FDA’s directive updates its April 2020 recommendations on the subject. It cites concerns about cleaning fixed endcap duodenoscopes and the increasing availability of models that eliminate the need for reprocessing.

The announcement highlighted the potential for a dramatic difference in between-patient contamination risk, reducing it “by half or more as compared to reusable, or fixed endcaps.”

“Interim results from one duodenoscope model with a removable component show a contamination rate of just 0.5%, as compared to older duodenoscope models which had contamination rates as high as 6%,” the FDA writes.

Duodenoscopes are used in more than 500,000 procedures each year in the United States and are key in assessing and treating diseases and conditions of the pancreas and bile ducts.
 

Upgrade to new models to decrease infections

Manufacturers no longer market fixed endcap models in the United States, but some health care facilities continue to use them. The FDA recommends that all fixed endcap models be replaced.

The FDA says some manufacturers are offering replacement programs to upgrade to a model with a disposable component at no cost.

Two fully disposable models and five with disposable components have been cleared by the FDA. (One model is no longer marketed and thus not listed here.)

Fully Disposable:

Ambu Innovation GmbH, Duodenoscope model aScope Duodeno

Boston Scientific Corporation, EXALT Model D Single-Use Duodenoscope

Disposable Components:

Fujifilm Corporation, Duodenoscope model ED-580XT 

Olympus Medical Systems, Evis Exera III Duodenovideoscope Olympus TJF-Q190V 

Pentax Medical, Duodenoscope model ED34-i10T2 

Pentax Medical, Duodenoscope model ED32-i10 

Additionally, the failure to correctly reprocess a duodenoscope could result in tissue or fluid from one patient transferring to a subsequent patient.

“In rare cases, this can lead to patient-to-patient disease transmission,” the FDA says.
 

Postmarket surveillance studies

In 2015, the FDA ordered three manufacturers of reusable devices (FujifilmOlympus, and Pentax) to conduct postmarket surveillance studies to determine contamination rates after reprocessing.

In 2019, the FDA also ordered postmarket surveillance studies to the makers of duodenoscopes with disposable endcaps to verify that the new designs reduce the contamination rate.

The final results of the fixed endcap design indicate that contamination rates were as high as 6.6% with high-concern organisms after contamination. High-concern organisms are those more often associated with disease, such as E coli and Pseudomonas contamination.

“As a result, Pentax and Olympus are withdrawing their fixed endcap duodenoscopes from the market, and Fujifilm has completed withdrawal of its fixed endcap duodenoscope,” the FDA writes.

Studies are not yet complete for duodenoscopes with removable components. As of August 12, 2021, the Fujifilm ED-580XT duodenoscope with a removable cap had 57% of the samples required. Interim results indicate that no samples tested positive for enough low-concern organisms to indicate a reprocessing failure, and only 0.5% tested positive for high-concern organisms.

In addition to the contamination risk sampling, each manufacturer was ordered to do postmarket surveillance studies to evaluate whether staff could understand and follow the manufacturer’s reprocessing instructions in real-world health care settings.

According to the FDA, the results showed that users frequently had difficulty understanding and following the manufacturers’ instructions and were not able to successfully complete reprocessing with the older models.

However, the newer models had high user success rates for understanding instructions and correctly performing reprocessing tasks, the FDA says.

A version of this article first appeared on Medscape.com.

Publications
Topics
Sections

Health care facilities and providers should complete the transition to fully disposable duodenoscopes and those with disposable components, the U.S. Food and Drug Administration announced this week after an analysis of postmarket surveillance studies was completed.

The FDA’s directive updates its April 2020 recommendations on the subject. It cites concerns about cleaning fixed endcap duodenoscopes and the increasing availability of models that eliminate the need for reprocessing.

The announcement highlighted the potential for a dramatic difference in between-patient contamination risk, reducing it “by half or more as compared to reusable, or fixed endcaps.”

“Interim results from one duodenoscope model with a removable component show a contamination rate of just 0.5%, as compared to older duodenoscope models which had contamination rates as high as 6%,” the FDA writes.

Duodenoscopes are used in more than 500,000 procedures each year in the United States and are key in assessing and treating diseases and conditions of the pancreas and bile ducts.
 

Upgrade to new models to decrease infections

Manufacturers no longer market fixed endcap models in the United States, but some health care facilities continue to use them. The FDA recommends that all fixed endcap models be replaced.

The FDA says some manufacturers are offering replacement programs to upgrade to a model with a disposable component at no cost.

Two fully disposable models and five with disposable components have been cleared by the FDA. (One model is no longer marketed and thus not listed here.)

Fully Disposable:

Ambu Innovation GmbH, Duodenoscope model aScope Duodeno

Boston Scientific Corporation, EXALT Model D Single-Use Duodenoscope

Disposable Components:

Fujifilm Corporation, Duodenoscope model ED-580XT 

Olympus Medical Systems, Evis Exera III Duodenovideoscope Olympus TJF-Q190V 

Pentax Medical, Duodenoscope model ED34-i10T2 

Pentax Medical, Duodenoscope model ED32-i10 

Additionally, the failure to correctly reprocess a duodenoscope could result in tissue or fluid from one patient transferring to a subsequent patient.

“In rare cases, this can lead to patient-to-patient disease transmission,” the FDA says.
 

Postmarket surveillance studies

In 2015, the FDA ordered three manufacturers of reusable devices (FujifilmOlympus, and Pentax) to conduct postmarket surveillance studies to determine contamination rates after reprocessing.

In 2019, the FDA also ordered postmarket surveillance studies to the makers of duodenoscopes with disposable endcaps to verify that the new designs reduce the contamination rate.

The final results of the fixed endcap design indicate that contamination rates were as high as 6.6% with high-concern organisms after contamination. High-concern organisms are those more often associated with disease, such as E coli and Pseudomonas contamination.

“As a result, Pentax and Olympus are withdrawing their fixed endcap duodenoscopes from the market, and Fujifilm has completed withdrawal of its fixed endcap duodenoscope,” the FDA writes.

Studies are not yet complete for duodenoscopes with removable components. As of August 12, 2021, the Fujifilm ED-580XT duodenoscope with a removable cap had 57% of the samples required. Interim results indicate that no samples tested positive for enough low-concern organisms to indicate a reprocessing failure, and only 0.5% tested positive for high-concern organisms.

In addition to the contamination risk sampling, each manufacturer was ordered to do postmarket surveillance studies to evaluate whether staff could understand and follow the manufacturer’s reprocessing instructions in real-world health care settings.

According to the FDA, the results showed that users frequently had difficulty understanding and following the manufacturers’ instructions and were not able to successfully complete reprocessing with the older models.

However, the newer models had high user success rates for understanding instructions and correctly performing reprocessing tasks, the FDA says.

A version of this article first appeared on Medscape.com.

Health care facilities and providers should complete the transition to fully disposable duodenoscopes and those with disposable components, the U.S. Food and Drug Administration announced this week after an analysis of postmarket surveillance studies was completed.

The FDA’s directive updates its April 2020 recommendations on the subject. It cites concerns about cleaning fixed endcap duodenoscopes and the increasing availability of models that eliminate the need for reprocessing.

The announcement highlighted the potential for a dramatic difference in between-patient contamination risk, reducing it “by half or more as compared to reusable, or fixed endcaps.”

“Interim results from one duodenoscope model with a removable component show a contamination rate of just 0.5%, as compared to older duodenoscope models which had contamination rates as high as 6%,” the FDA writes.

Duodenoscopes are used in more than 500,000 procedures each year in the United States and are key in assessing and treating diseases and conditions of the pancreas and bile ducts.
 

Upgrade to new models to decrease infections

Manufacturers no longer market fixed endcap models in the United States, but some health care facilities continue to use them. The FDA recommends that all fixed endcap models be replaced.

The FDA says some manufacturers are offering replacement programs to upgrade to a model with a disposable component at no cost.

Two fully disposable models and five with disposable components have been cleared by the FDA. (One model is no longer marketed and thus not listed here.)

Fully Disposable:

Ambu Innovation GmbH, Duodenoscope model aScope Duodeno

Boston Scientific Corporation, EXALT Model D Single-Use Duodenoscope

Disposable Components:

Fujifilm Corporation, Duodenoscope model ED-580XT 

Olympus Medical Systems, Evis Exera III Duodenovideoscope Olympus TJF-Q190V 

Pentax Medical, Duodenoscope model ED34-i10T2 

Pentax Medical, Duodenoscope model ED32-i10 

Additionally, the failure to correctly reprocess a duodenoscope could result in tissue or fluid from one patient transferring to a subsequent patient.

“In rare cases, this can lead to patient-to-patient disease transmission,” the FDA says.
 

Postmarket surveillance studies

In 2015, the FDA ordered three manufacturers of reusable devices (FujifilmOlympus, and Pentax) to conduct postmarket surveillance studies to determine contamination rates after reprocessing.

In 2019, the FDA also ordered postmarket surveillance studies to the makers of duodenoscopes with disposable endcaps to verify that the new designs reduce the contamination rate.

The final results of the fixed endcap design indicate that contamination rates were as high as 6.6% with high-concern organisms after contamination. High-concern organisms are those more often associated with disease, such as E coli and Pseudomonas contamination.

“As a result, Pentax and Olympus are withdrawing their fixed endcap duodenoscopes from the market, and Fujifilm has completed withdrawal of its fixed endcap duodenoscope,” the FDA writes.

Studies are not yet complete for duodenoscopes with removable components. As of August 12, 2021, the Fujifilm ED-580XT duodenoscope with a removable cap had 57% of the samples required. Interim results indicate that no samples tested positive for enough low-concern organisms to indicate a reprocessing failure, and only 0.5% tested positive for high-concern organisms.

In addition to the contamination risk sampling, each manufacturer was ordered to do postmarket surveillance studies to evaluate whether staff could understand and follow the manufacturer’s reprocessing instructions in real-world health care settings.

According to the FDA, the results showed that users frequently had difficulty understanding and following the manufacturers’ instructions and were not able to successfully complete reprocessing with the older models.

However, the newer models had high user success rates for understanding instructions and correctly performing reprocessing tasks, the FDA says.

A version of this article first appeared on Medscape.com.

Publications
Publications
Topics
Article Type
Sections
Disallow All Ads
Content Gating
No Gating (article Unlocked/Free)
Alternative CME
Disqus Comments
Default
Use ProPublica
Hide sidebar & use full width
render the right sidebar.
Conference Recap Checkbox
Not Conference Recap
Clinical Edge
Display the Slideshow in this Article
Medscape Article
Display survey writer
Reuters content
Disable Inline Native ads
WebMD Article