The relative safety of instrumental rotations in the second stage of labor remains controversial. Older reports suggest an unacceptable risk of fetal injury, while recent studies demonstrate more favorable outcomes without significant fetal or maternal morbidity. This study by Aiken and colleagues goes one step further by using propensity analysis to adjust for the likelihood of receiving an attempted instrumental rotation.

Details of the study
With a cohort of 833 women with second-stage positional abnormalities, Aiken and colleagues compared maternal and newborn outcomes associated with cesarean delivery (n = 534) with those of an attempted rotational procedure (n = 334). Among the attempted instrumental rotations, 299 (90%) were successful. By intention to treat, failed attempts at rotation and vaginal delivery were included in the instrumental rotation group. The authors relied on propensity analy­sis to adjust for selection bias.

Strengths and weaknesses
The main strengths of this study are the relatively large sample size, the inclusion of failed procedures in the forceps group based on intention to treat, the robust approach to adjusting for the likelihood of undergoing an attempted rotation, and the contemporary nature of the cohort.

However, the study has 4 important limitations:

  • More than 30% of rotations were attempted with vacuum devices. Many clinicians, including me, eschew vacuum deliveries for rotation due to reported higher failure rates and more scalp lacerations or other trauma. The analysis was not stratified by whether the rotation was attempted with a vacuum or Kielland forceps.
  • Information about maternal pelvic features, critical in determining the safety of any operative vaginal delivery, was not included. When the pelvis has anthropoid features, such as more room in the posterior segment, rotation is not needed and may be counterproductive. Android features raise the likelihood of dangerous outlet obstruction and generally suggest the need for cesarean delivery.
  • As Aiken and colleagues note, manual rotations followed by instrumental delivery from an occiput anterior position were not included.
  • The study was not stratified by whether the abnormal position was occiput posterior (OP) or occiput transverse (OT). Although the degree of rotation is greater with OP position, operative vaginal delivery from OT can be far more challenging.

What this evidence means for practice
Although this study does have limitations, it adds to the increasing number of contemporary reports suggesting that instrumental rotational procedures are safe. Though it is not without challenges, training in rotational forceps should continue.
— William H. Barth Jr, MD

Share your thoughts! Send your Letter to the Editor to [email protected]. Please include your name and the city and state in which you practice.

References

Article PDF
Author and Disclosure Information

expert commentary

William H. Barth Jr, MD

Chief, Division of Maternal-Fetal Medicine, Massachusetts General Hospital, and Associate Professor of Obstetrics, Gynecology, and Reproductive Biology, Harvard Medical School, Boston, Massachusetts.

The author reports no financial relationships relevant to this article.

Issue
OBG Management - 27(5)
Publications
Topics
Page Number
45,46
Legacy Keywords
William H. Barth Jr MD, rotational forceps, Examining the Evidence, rotational instrumental delivery, cesarean delivery, delayed neonatal respiration, reported critical incidents, fetal arterial umbilical pH, blood loss, significant hemorrhage, second state of labor, fetal injury, maternal morbidity, second-stage positional abnormalities, vaginal delivery, vacuum device, scalp lacerations, Kielland forceps, occiput posterior, OP, occiput transvers, OT, degree of rotation,
Sections
Author and Disclosure Information

expert commentary

William H. Barth Jr, MD

Chief, Division of Maternal-Fetal Medicine, Massachusetts General Hospital, and Associate Professor of Obstetrics, Gynecology, and Reproductive Biology, Harvard Medical School, Boston, Massachusetts.

The author reports no financial relationships relevant to this article.

Author and Disclosure Information

expert commentary

William H. Barth Jr, MD

Chief, Division of Maternal-Fetal Medicine, Massachusetts General Hospital, and Associate Professor of Obstetrics, Gynecology, and Reproductive Biology, Harvard Medical School, Boston, Massachusetts.

The author reports no financial relationships relevant to this article.

Article PDF
Article PDF
Related Articles

The relative safety of instrumental rotations in the second stage of labor remains controversial. Older reports suggest an unacceptable risk of fetal injury, while recent studies demonstrate more favorable outcomes without significant fetal or maternal morbidity. This study by Aiken and colleagues goes one step further by using propensity analysis to adjust for the likelihood of receiving an attempted instrumental rotation.

Details of the study
With a cohort of 833 women with second-stage positional abnormalities, Aiken and colleagues compared maternal and newborn outcomes associated with cesarean delivery (n = 534) with those of an attempted rotational procedure (n = 334). Among the attempted instrumental rotations, 299 (90%) were successful. By intention to treat, failed attempts at rotation and vaginal delivery were included in the instrumental rotation group. The authors relied on propensity analy­sis to adjust for selection bias.

Strengths and weaknesses
The main strengths of this study are the relatively large sample size, the inclusion of failed procedures in the forceps group based on intention to treat, the robust approach to adjusting for the likelihood of undergoing an attempted rotation, and the contemporary nature of the cohort.

However, the study has 4 important limitations:

  • More than 30% of rotations were attempted with vacuum devices. Many clinicians, including me, eschew vacuum deliveries for rotation due to reported higher failure rates and more scalp lacerations or other trauma. The analysis was not stratified by whether the rotation was attempted with a vacuum or Kielland forceps.
  • Information about maternal pelvic features, critical in determining the safety of any operative vaginal delivery, was not included. When the pelvis has anthropoid features, such as more room in the posterior segment, rotation is not needed and may be counterproductive. Android features raise the likelihood of dangerous outlet obstruction and generally suggest the need for cesarean delivery.
  • As Aiken and colleagues note, manual rotations followed by instrumental delivery from an occiput anterior position were not included.
  • The study was not stratified by whether the abnormal position was occiput posterior (OP) or occiput transverse (OT). Although the degree of rotation is greater with OP position, operative vaginal delivery from OT can be far more challenging.

What this evidence means for practice
Although this study does have limitations, it adds to the increasing number of contemporary reports suggesting that instrumental rotational procedures are safe. Though it is not without challenges, training in rotational forceps should continue.
— William H. Barth Jr, MD

Share your thoughts! Send your Letter to the Editor to [email protected]. Please include your name and the city and state in which you practice.

The relative safety of instrumental rotations in the second stage of labor remains controversial. Older reports suggest an unacceptable risk of fetal injury, while recent studies demonstrate more favorable outcomes without significant fetal or maternal morbidity. This study by Aiken and colleagues goes one step further by using propensity analysis to adjust for the likelihood of receiving an attempted instrumental rotation.

Details of the study
With a cohort of 833 women with second-stage positional abnormalities, Aiken and colleagues compared maternal and newborn outcomes associated with cesarean delivery (n = 534) with those of an attempted rotational procedure (n = 334). Among the attempted instrumental rotations, 299 (90%) were successful. By intention to treat, failed attempts at rotation and vaginal delivery were included in the instrumental rotation group. The authors relied on propensity analy­sis to adjust for selection bias.

Strengths and weaknesses
The main strengths of this study are the relatively large sample size, the inclusion of failed procedures in the forceps group based on intention to treat, the robust approach to adjusting for the likelihood of undergoing an attempted rotation, and the contemporary nature of the cohort.

However, the study has 4 important limitations:

  • More than 30% of rotations were attempted with vacuum devices. Many clinicians, including me, eschew vacuum deliveries for rotation due to reported higher failure rates and more scalp lacerations or other trauma. The analysis was not stratified by whether the rotation was attempted with a vacuum or Kielland forceps.
  • Information about maternal pelvic features, critical in determining the safety of any operative vaginal delivery, was not included. When the pelvis has anthropoid features, such as more room in the posterior segment, rotation is not needed and may be counterproductive. Android features raise the likelihood of dangerous outlet obstruction and generally suggest the need for cesarean delivery.
  • As Aiken and colleagues note, manual rotations followed by instrumental delivery from an occiput anterior position were not included.
  • The study was not stratified by whether the abnormal position was occiput posterior (OP) or occiput transverse (OT). Although the degree of rotation is greater with OP position, operative vaginal delivery from OT can be far more challenging.

What this evidence means for practice
Although this study does have limitations, it adds to the increasing number of contemporary reports suggesting that instrumental rotational procedures are safe. Though it is not without challenges, training in rotational forceps should continue.
— William H. Barth Jr, MD

Share your thoughts! Send your Letter to the Editor to [email protected]. Please include your name and the city and state in which you practice.

References

References

Issue
OBG Management - 27(5)
Issue
OBG Management - 27(5)
Page Number
45,46
Page Number
45,46
Publications
Publications
Topics
Article Type
Display Headline
Is it time to revive rotational forceps?
Display Headline
Is it time to revive rotational forceps?
Legacy Keywords
William H. Barth Jr MD, rotational forceps, Examining the Evidence, rotational instrumental delivery, cesarean delivery, delayed neonatal respiration, reported critical incidents, fetal arterial umbilical pH, blood loss, significant hemorrhage, second state of labor, fetal injury, maternal morbidity, second-stage positional abnormalities, vaginal delivery, vacuum device, scalp lacerations, Kielland forceps, occiput posterior, OP, occiput transvers, OT, degree of rotation,
Legacy Keywords
William H. Barth Jr MD, rotational forceps, Examining the Evidence, rotational instrumental delivery, cesarean delivery, delayed neonatal respiration, reported critical incidents, fetal arterial umbilical pH, blood loss, significant hemorrhage, second state of labor, fetal injury, maternal morbidity, second-stage positional abnormalities, vaginal delivery, vacuum device, scalp lacerations, Kielland forceps, occiput posterior, OP, occiput transvers, OT, degree of rotation,
Sections
Article Source

PURLs Copyright

Inside the Article

Article PDF Media