When the Food and Drug Administration (FDA) approved Skyla, a 13.5-mg levonorgestrel-releasing intrauterine system (LNG-IUS), early this year, it increased the number of IUDs available in the United States by one-third. The new device joins the copper IUD (ParaGard) and the 52-mg LNG-IUS (Mirena) as options for women seeking long-acting intrauterine contraception. The new device and Mirena both are manufactured by Bayer HealthCare Pharmaceuticals.

Skyla releases 14 μg of levonorgestrel daily, an amount that decreases progressively to 5 μg/day after 3 years. The device comes packaged within a sterile inserter and is approved for 3 years of use.

Approval followed a multinational Phase 3 trial

The trial involved 1,432 women 18 to 35 years old, 556 (38.8%) of whom were nulliparous. The primary efficacy endpoint used to assess contraceptive reliability was the pregnancy rate, calculated as the Pearl Index (PI). The PI estimate for the first year of use, based on the five pregnancies that occurred after the onset of treatment and within 7 days after Skyla removal or expulsion, was 0.41, with a 95% upper confidence limit of 0.96. The cumulative 3-year pregnancy rate, based on 10 pregnancies and estimated by the Kaplan-Meier method, was 0.9 per 100 women or 0.9%, with a 95% upper confidence limit of 1.7%.1

As a point of reference, the 12-month pregnancy rate for Mirena was ≤0.2%, and it was 0.5% over 5 years.2 The one-year pregnancy rate for ParaGard is 0.6% with perfect use, 0.8% with typical use.3

Of the women using Skyla, 21.9% discontinued because of an adverse event. The most common adverse reactions (occurring in ≥10% of users) were:

  • changes in bleeding patterns
  • vulvovaginitis
  • abdominal or pelvic pain
  • acne or seborrhea
  • ovarian cyst
  • headache.1
Family planning expert Mitchell Creinin, MD, professor and chair of obstetrics and gynecology at the University of California, Davis, does not see any cause for concern about adverse events.

“These adverse events include anything that occurs during study. For 20% of sexually active women to have at least one episode of vulvovaginitis over 3 years is not a problem”—nor is an occasional twinge of abdominal pain, he adds.

Insertion and removal

Skyla should be inserted during the first 7 days of the menstrual cycle or immediately after a first-trimester abortion. No back-up contraception is needed if it is inserted at these times.

Postpartum insertions should take place a minimum of 6 weeks after delivery or after a second-trimester abortion to allow for complete uterine involution.

Skyla should be removed after 3 years of use, preferably during the menstrual cycle to allow for immediate insertion of a new device or initiation of an alternate method of contraception.

Bleeding patterns may change during early use

During the first 3 to 6 months of use, women may experience irregular periods and an increase in the number of bleeding days. Women also may have frequent spotting or light bleeding. Some women may experience heavy bleeding during this interval. Over time, the number of bleeding and spotting days is likely to decline, and there is a small chance that periods may stop altogether. In clinical trials, 6% of Skyla users developed amenorrhea by the end of the first year.1 By comparison, 20% of Mirena users developed amenorrhea by the end of the first year.2

The only IUD for nulliparas?

Although Skyla is the only IUD studied in a significant number of nulliparous women during Phase 3 studies, and its labeling states specifically that it is appropriate for use in this population, the other two IUDs are also safe for nulliparas.

“The label for ParaGard includes nulliparous and parous women,” notes Dr. Creinin. “The label for Mirena does not exclude nulliparous women, and there are huge numbers of studies about its use in teens and nulliparas. The only issue for Skyla is that the data presented to the FDA included nulliparous women, so the label reflects this fact.”

What’s the bottom line?

Skyla is an option for women who may desire a smaller IUD with a shorter duration of use. Although placement of Skyla is enhanced by a new inserter, compared with Mirena and ParaGard, Dr. Creinin notes that all IUD insertions, regardless of brand, are fairly straightforward. For example, a randomized trial by Gemzell-Danielsson and colleagues found that IUD insertion was “very difficult” in only 2 of 239 (0.8%) women receiving a prototype of Skyla and 4 of 254 (1.6%) women receiving Mirena.4

The cost of Skyla is approximately $650 (for 3 years of use) versus roughly $850 for Mirena and ParaGard (for 5 and 10 years of use, respectively).

 

 

In general, if a woman specifically requests a shorter duration of efficacy, then Skyla may be appropriate for her, although any IUD can be removed at any time if a patient decides to become pregnant.

We want to hear from you! Tell us what you think.

References

1. Skyla [package insert]. Wayne, NJ: Bayer HealthCare Pharmaceuticals; 2013.

2. Mirena [package insert]. Wayne, NJ: Bayer HealthCare Pharmaceuticals; 2013.

3. ParaGard [package insert]. Sellersville, PA: Teva Women’s Health; 2013.

4. Gemzell-Danielsson K, Schellschmidt I, Apter D. A randomized, phase II study describing the efficacy, bleeding profile, and safety of two low-dose levonorgestrel-releasing intrauterine contraceptive systems and Mirena. Fertil Steril. 2012;97(3):616-22.e1-e3.

More NEWS FOR YOUR PRACTICE…

Vitamin D in pregnancy and postmenopause: Recent evidence implies it may not be so importantNative tissue is superior to vaginal mesh for prolapse repair, two studies reportPostpartum anxiety more common than depressionRobotic surgery not the best for hysterectomy, ACOG saysRobotically assisted hysterectomy is on the rise for benign gynecologic disordersMaternal folic acid use linked to reduced autism riskSupport for biennial over annual mammography in older womenUse of emergency contraception has more than doubled

Author and Disclosure Information

Janelle Yates,
Senior Editor

Issue
OBG Management - 25(4)
Publications
Topics
Legacy Keywords
Janelle Yates;Mitchell Creinin MD;Skyla;IUD;Bayer HealthCare Pharmaceuticals;intrauterine device;mini levonorgestrel-releasing intrauterine system;LNG-IUS;bleeding;duration of efficacy;FDA;copper IUD;ParaGard;Mirena;long-acting intrauterine contraception;nulliparous;contraceptive reliability;expulsion;changes in bleeding patterns;vulvovaginitis;abdominal or pelvic pain;acne;seborrhea;ovarian cyst;headache;menstrual cycle;abortion;postpartum insertion;complete uterine involution;IUD insertion;
Sections
Author and Disclosure Information

Janelle Yates,
Senior Editor

Author and Disclosure Information

Janelle Yates,
Senior Editor

HAVE YOU READ THIS RELATED ARTICLE

New IUD approved by FDA (Web NEWS, January 2013)

When the Food and Drug Administration (FDA) approved Skyla, a 13.5-mg levonorgestrel-releasing intrauterine system (LNG-IUS), early this year, it increased the number of IUDs available in the United States by one-third. The new device joins the copper IUD (ParaGard) and the 52-mg LNG-IUS (Mirena) as options for women seeking long-acting intrauterine contraception. The new device and Mirena both are manufactured by Bayer HealthCare Pharmaceuticals.

Skyla releases 14 μg of levonorgestrel daily, an amount that decreases progressively to 5 μg/day after 3 years. The device comes packaged within a sterile inserter and is approved for 3 years of use.

Approval followed a multinational Phase 3 trial

The trial involved 1,432 women 18 to 35 years old, 556 (38.8%) of whom were nulliparous. The primary efficacy endpoint used to assess contraceptive reliability was the pregnancy rate, calculated as the Pearl Index (PI). The PI estimate for the first year of use, based on the five pregnancies that occurred after the onset of treatment and within 7 days after Skyla removal or expulsion, was 0.41, with a 95% upper confidence limit of 0.96. The cumulative 3-year pregnancy rate, based on 10 pregnancies and estimated by the Kaplan-Meier method, was 0.9 per 100 women or 0.9%, with a 95% upper confidence limit of 1.7%.1

As a point of reference, the 12-month pregnancy rate for Mirena was ≤0.2%, and it was 0.5% over 5 years.2 The one-year pregnancy rate for ParaGard is 0.6% with perfect use, 0.8% with typical use.3

Of the women using Skyla, 21.9% discontinued because of an adverse event. The most common adverse reactions (occurring in ≥10% of users) were:

  • changes in bleeding patterns
  • vulvovaginitis
  • abdominal or pelvic pain
  • acne or seborrhea
  • ovarian cyst
  • headache.1
Family planning expert Mitchell Creinin, MD, professor and chair of obstetrics and gynecology at the University of California, Davis, does not see any cause for concern about adverse events.

“These adverse events include anything that occurs during study. For 20% of sexually active women to have at least one episode of vulvovaginitis over 3 years is not a problem”—nor is an occasional twinge of abdominal pain, he adds.

Insertion and removal

Skyla should be inserted during the first 7 days of the menstrual cycle or immediately after a first-trimester abortion. No back-up contraception is needed if it is inserted at these times.

Postpartum insertions should take place a minimum of 6 weeks after delivery or after a second-trimester abortion to allow for complete uterine involution.

Skyla should be removed after 3 years of use, preferably during the menstrual cycle to allow for immediate insertion of a new device or initiation of an alternate method of contraception.

Bleeding patterns may change during early use

During the first 3 to 6 months of use, women may experience irregular periods and an increase in the number of bleeding days. Women also may have frequent spotting or light bleeding. Some women may experience heavy bleeding during this interval. Over time, the number of bleeding and spotting days is likely to decline, and there is a small chance that periods may stop altogether. In clinical trials, 6% of Skyla users developed amenorrhea by the end of the first year.1 By comparison, 20% of Mirena users developed amenorrhea by the end of the first year.2

The only IUD for nulliparas?

Although Skyla is the only IUD studied in a significant number of nulliparous women during Phase 3 studies, and its labeling states specifically that it is appropriate for use in this population, the other two IUDs are also safe for nulliparas.

“The label for ParaGard includes nulliparous and parous women,” notes Dr. Creinin. “The label for Mirena does not exclude nulliparous women, and there are huge numbers of studies about its use in teens and nulliparas. The only issue for Skyla is that the data presented to the FDA included nulliparous women, so the label reflects this fact.”

What’s the bottom line?

Skyla is an option for women who may desire a smaller IUD with a shorter duration of use. Although placement of Skyla is enhanced by a new inserter, compared with Mirena and ParaGard, Dr. Creinin notes that all IUD insertions, regardless of brand, are fairly straightforward. For example, a randomized trial by Gemzell-Danielsson and colleagues found that IUD insertion was “very difficult” in only 2 of 239 (0.8%) women receiving a prototype of Skyla and 4 of 254 (1.6%) women receiving Mirena.4

The cost of Skyla is approximately $650 (for 3 years of use) versus roughly $850 for Mirena and ParaGard (for 5 and 10 years of use, respectively).

 

 

In general, if a woman specifically requests a shorter duration of efficacy, then Skyla may be appropriate for her, although any IUD can be removed at any time if a patient decides to become pregnant.

We want to hear from you! Tell us what you think.

HAVE YOU READ THIS RELATED ARTICLE

New IUD approved by FDA (Web NEWS, January 2013)

When the Food and Drug Administration (FDA) approved Skyla, a 13.5-mg levonorgestrel-releasing intrauterine system (LNG-IUS), early this year, it increased the number of IUDs available in the United States by one-third. The new device joins the copper IUD (ParaGard) and the 52-mg LNG-IUS (Mirena) as options for women seeking long-acting intrauterine contraception. The new device and Mirena both are manufactured by Bayer HealthCare Pharmaceuticals.

Skyla releases 14 μg of levonorgestrel daily, an amount that decreases progressively to 5 μg/day after 3 years. The device comes packaged within a sterile inserter and is approved for 3 years of use.

Approval followed a multinational Phase 3 trial

The trial involved 1,432 women 18 to 35 years old, 556 (38.8%) of whom were nulliparous. The primary efficacy endpoint used to assess contraceptive reliability was the pregnancy rate, calculated as the Pearl Index (PI). The PI estimate for the first year of use, based on the five pregnancies that occurred after the onset of treatment and within 7 days after Skyla removal or expulsion, was 0.41, with a 95% upper confidence limit of 0.96. The cumulative 3-year pregnancy rate, based on 10 pregnancies and estimated by the Kaplan-Meier method, was 0.9 per 100 women or 0.9%, with a 95% upper confidence limit of 1.7%.1

As a point of reference, the 12-month pregnancy rate for Mirena was ≤0.2%, and it was 0.5% over 5 years.2 The one-year pregnancy rate for ParaGard is 0.6% with perfect use, 0.8% with typical use.3

Of the women using Skyla, 21.9% discontinued because of an adverse event. The most common adverse reactions (occurring in ≥10% of users) were:

  • changes in bleeding patterns
  • vulvovaginitis
  • abdominal or pelvic pain
  • acne or seborrhea
  • ovarian cyst
  • headache.1
Family planning expert Mitchell Creinin, MD, professor and chair of obstetrics and gynecology at the University of California, Davis, does not see any cause for concern about adverse events.

“These adverse events include anything that occurs during study. For 20% of sexually active women to have at least one episode of vulvovaginitis over 3 years is not a problem”—nor is an occasional twinge of abdominal pain, he adds.

Insertion and removal

Skyla should be inserted during the first 7 days of the menstrual cycle or immediately after a first-trimester abortion. No back-up contraception is needed if it is inserted at these times.

Postpartum insertions should take place a minimum of 6 weeks after delivery or after a second-trimester abortion to allow for complete uterine involution.

Skyla should be removed after 3 years of use, preferably during the menstrual cycle to allow for immediate insertion of a new device or initiation of an alternate method of contraception.

Bleeding patterns may change during early use

During the first 3 to 6 months of use, women may experience irregular periods and an increase in the number of bleeding days. Women also may have frequent spotting or light bleeding. Some women may experience heavy bleeding during this interval. Over time, the number of bleeding and spotting days is likely to decline, and there is a small chance that periods may stop altogether. In clinical trials, 6% of Skyla users developed amenorrhea by the end of the first year.1 By comparison, 20% of Mirena users developed amenorrhea by the end of the first year.2

The only IUD for nulliparas?

Although Skyla is the only IUD studied in a significant number of nulliparous women during Phase 3 studies, and its labeling states specifically that it is appropriate for use in this population, the other two IUDs are also safe for nulliparas.

“The label for ParaGard includes nulliparous and parous women,” notes Dr. Creinin. “The label for Mirena does not exclude nulliparous women, and there are huge numbers of studies about its use in teens and nulliparas. The only issue for Skyla is that the data presented to the FDA included nulliparous women, so the label reflects this fact.”

What’s the bottom line?

Skyla is an option for women who may desire a smaller IUD with a shorter duration of use. Although placement of Skyla is enhanced by a new inserter, compared with Mirena and ParaGard, Dr. Creinin notes that all IUD insertions, regardless of brand, are fairly straightforward. For example, a randomized trial by Gemzell-Danielsson and colleagues found that IUD insertion was “very difficult” in only 2 of 239 (0.8%) women receiving a prototype of Skyla and 4 of 254 (1.6%) women receiving Mirena.4

The cost of Skyla is approximately $650 (for 3 years of use) versus roughly $850 for Mirena and ParaGard (for 5 and 10 years of use, respectively).

 

 

In general, if a woman specifically requests a shorter duration of efficacy, then Skyla may be appropriate for her, although any IUD can be removed at any time if a patient decides to become pregnant.

We want to hear from you! Tell us what you think.

References

1. Skyla [package insert]. Wayne, NJ: Bayer HealthCare Pharmaceuticals; 2013.

2. Mirena [package insert]. Wayne, NJ: Bayer HealthCare Pharmaceuticals; 2013.

3. ParaGard [package insert]. Sellersville, PA: Teva Women’s Health; 2013.

4. Gemzell-Danielsson K, Schellschmidt I, Apter D. A randomized, phase II study describing the efficacy, bleeding profile, and safety of two low-dose levonorgestrel-releasing intrauterine contraceptive systems and Mirena. Fertil Steril. 2012;97(3):616-22.e1-e3.

More NEWS FOR YOUR PRACTICE…

Vitamin D in pregnancy and postmenopause: Recent evidence implies it may not be so importantNative tissue is superior to vaginal mesh for prolapse repair, two studies reportPostpartum anxiety more common than depressionRobotic surgery not the best for hysterectomy, ACOG saysRobotically assisted hysterectomy is on the rise for benign gynecologic disordersMaternal folic acid use linked to reduced autism riskSupport for biennial over annual mammography in older womenUse of emergency contraception has more than doubled

References

1. Skyla [package insert]. Wayne, NJ: Bayer HealthCare Pharmaceuticals; 2013.

2. Mirena [package insert]. Wayne, NJ: Bayer HealthCare Pharmaceuticals; 2013.

3. ParaGard [package insert]. Sellersville, PA: Teva Women’s Health; 2013.

4. Gemzell-Danielsson K, Schellschmidt I, Apter D. A randomized, phase II study describing the efficacy, bleeding profile, and safety of two low-dose levonorgestrel-releasing intrauterine contraceptive systems and Mirena. Fertil Steril. 2012;97(3):616-22.e1-e3.

More NEWS FOR YOUR PRACTICE…

Vitamin D in pregnancy and postmenopause: Recent evidence implies it may not be so importantNative tissue is superior to vaginal mesh for prolapse repair, two studies reportPostpartum anxiety more common than depressionRobotic surgery not the best for hysterectomy, ACOG saysRobotically assisted hysterectomy is on the rise for benign gynecologic disordersMaternal folic acid use linked to reduced autism riskSupport for biennial over annual mammography in older womenUse of emergency contraception has more than doubled

Issue
OBG Management - 25(4)
Issue
OBG Management - 25(4)
Publications
Publications
Topics
Article Type
Display Headline
The newly approved IUD: Which patients is Skyla appropriate for?
Display Headline
The newly approved IUD: Which patients is Skyla appropriate for?
Legacy Keywords
Janelle Yates;Mitchell Creinin MD;Skyla;IUD;Bayer HealthCare Pharmaceuticals;intrauterine device;mini levonorgestrel-releasing intrauterine system;LNG-IUS;bleeding;duration of efficacy;FDA;copper IUD;ParaGard;Mirena;long-acting intrauterine contraception;nulliparous;contraceptive reliability;expulsion;changes in bleeding patterns;vulvovaginitis;abdominal or pelvic pain;acne;seborrhea;ovarian cyst;headache;menstrual cycle;abortion;postpartum insertion;complete uterine involution;IUD insertion;
Legacy Keywords
Janelle Yates;Mitchell Creinin MD;Skyla;IUD;Bayer HealthCare Pharmaceuticals;intrauterine device;mini levonorgestrel-releasing intrauterine system;LNG-IUS;bleeding;duration of efficacy;FDA;copper IUD;ParaGard;Mirena;long-acting intrauterine contraception;nulliparous;contraceptive reliability;expulsion;changes in bleeding patterns;vulvovaginitis;abdominal or pelvic pain;acne;seborrhea;ovarian cyst;headache;menstrual cycle;abortion;postpartum insertion;complete uterine involution;IUD insertion;
Sections
Article Source

PURLs Copyright

Inside the Article