WASHINGTON — A single-use bacteriophage amplification test kit was able to both accurately identify Staphylococcus aureus and determine whether it was methicillin sensitive or resistant within 5 hours in a study of clinical bacteremia isolates.
The findings suggest that it is possible to slash the diagnostic time for bacteremia—from 2–3 days to 5 hours—and obtain rapid results that will guide treatment and prevent overuse of broad-spectrum antibiotics, Dr. J. Drew Smith said in an interview during his poster presentation at the jointly held annual Interscience Conference on Antimicrobial Agents and Chemotherapy and the annual meeting of the Infectious Diseases Society of America.
The test, made by MicroPhage Inc., uses bacteriophage amplification technology, which detects proteins produced by viruses that are selected to amplify in response to S. aureus. Blood culture samples are mixed in two separate tubes and placed in an incubator for 5 hours. The tubes are removed and six drops of each sample are applied to dipstick-type detectors similar to those used in home pregnancy tests. One tube determines whether or not the sample contains S. aureus; the other determines whether the bacteria are antibiotic resistant or susceptible, said Dr. Smith, director of research and development at MicroPhage.
In a panel of 120 S. aureus clinical isolates and 120 closely related nonpathogenic coagulase-negative staphylococci, the identity test for S. aureus had a sensitivity of 93% and a specificity of 96%. Among the strains identified as S. aureus, methicillin susceptibility was determined with 99% sensitivity and 99% specificity. Only 1.8% of samples were falsely identified as methicillin-resistant S. aureus (MRSA) and no samples were falsely identified as methicillin sensitive (MSSA), Dr. Smith and his associates reported.
Current polymerase chain reaction (PCR) technology allows for rapid detection of MRSA but doesn't accurately determine susceptibility. With the bacteriophage test, a result indicating MSSA allows for the patient to be safely switched from empiric vancomycin to nafcillin or another conventional β-lactam antibiotic, which are more effective against S. aureus than is vancomycin and can reduce mortality by 30%–50% if the organism is susceptible. The PCR test gives too many false positives for MSSA in order to be used for this purpose, Dr. Smith explained in the interview.
Bacteriophage amplification also could be used to screen patients for MRSA carriage. In a separate study presented in another poster, nasal swabs collected from preoperative and ICU patients were streaked on agar plates for MRSA detection. The swabs were transferred to MicroPhage tubes, incubated for 7–24 hours, and read in the same way as was done for the bacteremia test. This time, 32 samples were read at 7 and 24 hours and 77 were read at 12–18 hours and again at 18–24 hours. (More time is needed for nasal swabs than blood cultures because fewer bacteria are present, Dr. Smith explained.)
Sensitivity for detecting MRSA nasal carriage was 33% at 7 hours, 92% at 12–18 hours, and 100% at 18–24 hours. Specificity was 100% at 7 hours and 98% at 12–18 and 18–24 hours. Positive predictive value was 100% at 7 hours and 88% by 18–24 hours; negative predictive value rose from 94% at 7 hours to 100% at 18–24 hours.
Lab personnel were trained to use the test in less than half an hour, and it required no specialized or dedicated equipment. Moreover, “the test is flexible with respect to read times, allowing it to be adapted to a variety of testing and reporting schedules,” the investigators said.
MicroPhage is hoping to market both uses for the technology to community hospitals and to offer the nasal tests to outpatient settings such as nursing homes or surgicenters. Clinical testing will begin in early 2009, and the company hopes to obtain licensure by early 2010, Dr. Smith said.
WASHINGTON — A single-use bacteriophage amplification test kit was able to both accurately identify Staphylococcus aureus and determine whether it was methicillin sensitive or resistant within 5 hours in a study of clinical bacteremia isolates.
The findings suggest that it is possible to slash the diagnostic time for bacteremia—from 2–3 days to 5 hours—and obtain rapid results that will guide treatment and prevent overuse of broad-spectrum antibiotics, Dr. J. Drew Smith said in an interview during his poster presentation at the jointly held annual Interscience Conference on Antimicrobial Agents and Chemotherapy and the annual meeting of the Infectious Diseases Society of America.
The test, made by MicroPhage Inc., uses bacteriophage amplification technology, which detects proteins produced by viruses that are selected to amplify in response to S. aureus. Blood culture samples are mixed in two separate tubes and placed in an incubator for 5 hours. The tubes are removed and six drops of each sample are applied to dipstick-type detectors similar to those used in home pregnancy tests. One tube determines whether or not the sample contains S. aureus; the other determines whether the bacteria are antibiotic resistant or susceptible, said Dr. Smith, director of research and development at MicroPhage.
In a panel of 120 S. aureus clinical isolates and 120 closely related nonpathogenic coagulase-negative staphylococci, the identity test for S. aureus had a sensitivity of 93% and a specificity of 96%. Among the strains identified as S. aureus, methicillin susceptibility was determined with 99% sensitivity and 99% specificity. Only 1.8% of samples were falsely identified as methicillin-resistant S. aureus (MRSA) and no samples were falsely identified as methicillin sensitive (MSSA), Dr. Smith and his associates reported.
Current polymerase chain reaction (PCR) technology allows for rapid detection of MRSA but doesn't accurately determine susceptibility. With the bacteriophage test, a result indicating MSSA allows for the patient to be safely switched from empiric vancomycin to nafcillin or another conventional β-lactam antibiotic, which are more effective against S. aureus than is vancomycin and can reduce mortality by 30%–50% if the organism is susceptible. The PCR test gives too many false positives for MSSA in order to be used for this purpose, Dr. Smith explained in the interview.
Bacteriophage amplification also could be used to screen patients for MRSA carriage. In a separate study presented in another poster, nasal swabs collected from preoperative and ICU patients were streaked on agar plates for MRSA detection. The swabs were transferred to MicroPhage tubes, incubated for 7–24 hours, and read in the same way as was done for the bacteremia test. This time, 32 samples were read at 7 and 24 hours and 77 were read at 12–18 hours and again at 18–24 hours. (More time is needed for nasal swabs than blood cultures because fewer bacteria are present, Dr. Smith explained.)
Sensitivity for detecting MRSA nasal carriage was 33% at 7 hours, 92% at 12–18 hours, and 100% at 18–24 hours. Specificity was 100% at 7 hours and 98% at 12–18 and 18–24 hours. Positive predictive value was 100% at 7 hours and 88% by 18–24 hours; negative predictive value rose from 94% at 7 hours to 100% at 18–24 hours.
Lab personnel were trained to use the test in less than half an hour, and it required no specialized or dedicated equipment. Moreover, “the test is flexible with respect to read times, allowing it to be adapted to a variety of testing and reporting schedules,” the investigators said.
MicroPhage is hoping to market both uses for the technology to community hospitals and to offer the nasal tests to outpatient settings such as nursing homes or surgicenters. Clinical testing will begin in early 2009, and the company hopes to obtain licensure by early 2010, Dr. Smith said.
WASHINGTON — A single-use bacteriophage amplification test kit was able to both accurately identify Staphylococcus aureus and determine whether it was methicillin sensitive or resistant within 5 hours in a study of clinical bacteremia isolates.
The findings suggest that it is possible to slash the diagnostic time for bacteremia—from 2–3 days to 5 hours—and obtain rapid results that will guide treatment and prevent overuse of broad-spectrum antibiotics, Dr. J. Drew Smith said in an interview during his poster presentation at the jointly held annual Interscience Conference on Antimicrobial Agents and Chemotherapy and the annual meeting of the Infectious Diseases Society of America.
The test, made by MicroPhage Inc., uses bacteriophage amplification technology, which detects proteins produced by viruses that are selected to amplify in response to S. aureus. Blood culture samples are mixed in two separate tubes and placed in an incubator for 5 hours. The tubes are removed and six drops of each sample are applied to dipstick-type detectors similar to those used in home pregnancy tests. One tube determines whether or not the sample contains S. aureus; the other determines whether the bacteria are antibiotic resistant or susceptible, said Dr. Smith, director of research and development at MicroPhage.
In a panel of 120 S. aureus clinical isolates and 120 closely related nonpathogenic coagulase-negative staphylococci, the identity test for S. aureus had a sensitivity of 93% and a specificity of 96%. Among the strains identified as S. aureus, methicillin susceptibility was determined with 99% sensitivity and 99% specificity. Only 1.8% of samples were falsely identified as methicillin-resistant S. aureus (MRSA) and no samples were falsely identified as methicillin sensitive (MSSA), Dr. Smith and his associates reported.
Current polymerase chain reaction (PCR) technology allows for rapid detection of MRSA but doesn't accurately determine susceptibility. With the bacteriophage test, a result indicating MSSA allows for the patient to be safely switched from empiric vancomycin to nafcillin or another conventional β-lactam antibiotic, which are more effective against S. aureus than is vancomycin and can reduce mortality by 30%–50% if the organism is susceptible. The PCR test gives too many false positives for MSSA in order to be used for this purpose, Dr. Smith explained in the interview.
Bacteriophage amplification also could be used to screen patients for MRSA carriage. In a separate study presented in another poster, nasal swabs collected from preoperative and ICU patients were streaked on agar plates for MRSA detection. The swabs were transferred to MicroPhage tubes, incubated for 7–24 hours, and read in the same way as was done for the bacteremia test. This time, 32 samples were read at 7 and 24 hours and 77 were read at 12–18 hours and again at 18–24 hours. (More time is needed for nasal swabs than blood cultures because fewer bacteria are present, Dr. Smith explained.)
Sensitivity for detecting MRSA nasal carriage was 33% at 7 hours, 92% at 12–18 hours, and 100% at 18–24 hours. Specificity was 100% at 7 hours and 98% at 12–18 and 18–24 hours. Positive predictive value was 100% at 7 hours and 88% by 18–24 hours; negative predictive value rose from 94% at 7 hours to 100% at 18–24 hours.
Lab personnel were trained to use the test in less than half an hour, and it required no specialized or dedicated equipment. Moreover, “the test is flexible with respect to read times, allowing it to be adapted to a variety of testing and reporting schedules,” the investigators said.