Systematic Review of Physiologic Monitor Alarm Characteristics and Pragmatic Interventions to Reduce Alarm Frequency

Clinical alarm safety has become a recent target for improvement in many hospitals. In 2013, The Joint Commission released a National Patient Safety Goal prompting accredited hospitals to establish alarm safety as a hospital priority, identify the most important alarm signals to manage, and, by 2016, develop policies and procedures that address alarm management.[1] In addition, the Emergency Care Research Institute has named alarm hazards the top health technology hazard each year since 2012.[2]

The primary arguments supporting the elevation of alarm management to a national hospital priority in the United States include the following: (1) clinicians rely on alarms to notify them of important physiologic changes, (2) alarms occur frequently and usually do not warrant clinical intervention, and (3) alarm overload renders clinicians unable to respond to all alarms, resulting in alarm fatigue: responding more slowly or ignoring alarms that may represent actual clinical deterioration.[3, 4] These arguments are built largely on anecdotal data, reported safety event databases, and small studies that have not previously been systematically analyzed.

Despite the national focus on alarms, we still know very little about fundamental questions key to improving alarm safety. In this systematic review, we aimed to answer 3 key questions about physiologic monitor alarms: (1) What proportion of alarms warrant attention or clinical intervention (ie, actionable alarms), and how does this proportion vary between adult and pediatric populations and between intensive care unit (ICU) and ward settings? (2) What is the relationship between alarm exposure and clinician response time? (3) What interventions are effective in reducing the frequency of alarms?

We limited our scope to monitor alarms because few studies have evaluated the characteristics of alarms from other medical devices, and because missing relevant monitor alarms could adversely impact patient safety.

METHODS

We performed a systematic review of the literature in accordance with the Meta‐Analysis of Observational Studies in Epidemiology guidelines[5] and developed this manuscript using the Preferred Reporting Items for Systematic Reviews and Meta‐Analyses (PRISMA) statement.[6]

Eligibility Criteria

With help from an experienced biomedical librarian (C.D.S.), we searched PubMed, the Cumulative Index to Nursing and Allied Health Literature, Scopus, Cochrane Library, ClinicalTrials.gov, and Google Scholar from January 1980 through April 2015 (see Supporting Information in the online version of this article for the search terms and queries). We hand searched the reference lists of included articles and reviewed our personal libraries to identify additional relevant studies.

We included peer‐reviewed, original research studies published in English, Spanish, or French that addressed the questions outlined above. Eligible patient populations were children and adults admitted to hospital inpatient units and emergency departments (EDs). We excluded alarms in procedural suites or operating rooms (typically responded to by anesthesiologists already with the patient) because of the differences in environment of care, staff‐to‐patient ratio, and equipment. We included observational studies reporting the actionability of physiologic monitor alarms (ie, alarms warranting special attention or clinical intervention), as well as nurse responses to these alarms. We excluded studies focused on the effects of alarms unrelated to patient safety, such as families' and patients' stress, noise, or sleep disturbance. We included only intervention studies evaluating pragmatic interventions ready for clinical implementation (ie, not experimental devices or software algorithms).

Selection Process and Data Extraction

First, 2 authors screened the titles and abstracts of articles for eligibility. To maximize sensitivity, if at least 1 author considered the article relevant, the article proceeded to full‐text review. Second, the full texts of articles screened were independently reviewed by 2 authors in an unblinded fashion to determine their eligibility. Any disagreements concerning eligibility were resolved by team consensus. To assure consistency in eligibility determinations across the team, a core group of the authors (C.W.P, C.P.B., E.E., and V.V.G.) held a series of meetings to review and discuss each potentially eligible article and reach consensus on the final list of included articles. Two authors independently extracted the following characteristics from included studies: alarm review methods, analytic design, fidelity measurement, consideration of unintended adverse safety consequences, and key results. Reviewers were not blinded to journal, authors, or affiliations.

Synthesis of Results and Risk Assessment

Given the high degree of heterogeneity in methodology, we were unable to generate summary proportions of the observational studies or perform a meta‐analysis of the intervention studies. Thus, we organized the studies into clinically relevant categories and presented key aspects in tables. Due to the heterogeneity of the studies and the controversy surrounding quality scores,[5] we did not generate summary scores of study quality. Instead, we evaluated and reported key design elements that had the potential to bias the results. To recognize the more comprehensive studies in the field, we developed by consensus a set of characteristics that distinguished studies with lower risk of bias. These characteristics are shown and defined in Table 1.

General Characteristics of Included Studies
First Author and Publication Year Alarm Review Method Indicators of Potential Bias for Observational Studies Indicators of Potential Bias for Intervention Studies
Monitor System Direct Observation Medical Record Review Rhythm Annotation Video Observation Remote Monitoring Staff Medical Device Industry Involved Two Independent Reviewers At Least 1 Reviewer Is a Clinical Expert Reviewer Not Simultaneously in Patient Care Clear Definition of Alarm Actionability Census Included Statistical Testing or QI SPC Methods Fidelity Assessed Safety Assessed Lower Risk of Bias
  • NOTE: Lower risk of bias for observational studies required all of the following characteristics be reported: (1) two independent reviewers of alarms, (2) at least 1 clinical expert reviewer (ie, physician or nurse), (3) the reviewer was not simultaneously involved in clinical care, and (4) there was a clear definition of alarm actionability provided in the article. These indicators assess detection bias, observer bias, analytical bias, and reporting bias and were derived from the Meta‐analysis of Observational Studies in Epidemiology checklist.[5] Lower risk of bias for intervention studies required all of the following characteristics be reported: (1) patient census accounted for in analysis, (2) formal statistical testing of effect or statistical process control methods used to evaluate effect, (3) intervention fidelity measured and reported (defined broadly as some measurement of whether the intervention was delivered as intended), and (4) relevant patient safety outcomes measured and reported (eg, the rate of code blue events before and after the intervention was implemented). These indicators assess reporting bias and internal validity bias and were derived from the Downs and Black checklist.[42] Monitor system: alarm data were electronically collected directly from the physiologic monitors and saved on a computer device through software such as BedMasterEx. Direct observation: an in‐person observer, such as a research assistant or a nurse, takes note of the alarm data and/or responses to alarms. Medical record review: data on alarms and/or responses to alarms were extracted from the patient medical records. Rhythm annotation: data on waveforms from cardiac monitors were collected and saved on a computer device through software such as BedMasterEx. Video observation: video cameras were set up in the patient's room and recorded data on alarms and/or responses to alarms. Remote monitor staff: clinicians situated at a remote location observe the patient via video camera and may be able to communicate with the patient or the patient's assigned nurse. Abbreviations: QI, quality improvement; RN, registered nurse; SPC, statistical process control. *Monitor system + RN interrogation. Assigned nurse making observations. Monitor from central station. Alarm outcome reported using run chart, and fidelity outcomes presented using statistical process control charts.

Adult Observational
Atzema 2006[7] ✓* ✓ ✓
Billinghurst 2003[8] ✓ ✓ ✓ ✓
Biot 2000[9] ✓ ✓ ✓ ✓
Chambrin 1999[10] ✓ ✓ ✓ ✓
Drew 2014[11] ✓ ✓ ✓ ✓ ✓ ✓ ✓ ✓
Gazarian 2014[12] ✓ ✓ ✓ ✓ ✓
Grges 2009[13] ✓ ✓ ✓ ✓
Gross 2011[15] ✓ ✓ ✓ ✓ ✓ ✓ ✓
Inokuchi 2013[14] ✓ ✓ ✓ ✓ ✓ ✓ ✓
Koski 1990[16] ✓ ✓ ✓ ✓
Morales Snchez 2014[17] ✓ ✓ ✓ ✓
Pergher 2014[18] ✓ ✓
Siebig 2010[19] ✓ ✓ ✓ ✓ ✓ ✓
Voepel‐Lewis 2013[20] ✓ ✓ ✓ ✓
Way 2014[21] ✓ ✓ ✓
Pediatric Observational
Bonafide 2015[22] ✓ ✓ ✓ ✓ ✓ ✓ ✓
Lawless 1994[23] ✓ ✓
Rosman 2013[24] ✓ ✓ ✓ ✓ ✓
Talley 2011[25] ✓ ✓ ✓ ✓ ✓
Tsien 1997[26] ✓ ✓ ✓ ✓
van Pul 2015[27] ✓
Varpio 2012[28] ✓ ✓ ✓ ✓
Mixed Adult and Pediatric Observational
O'Carroll 1986[29] ✓
Wiklund 1994[30] ✓ ✓ ✓ ✓
Adult Intervention
Albert 2015[32] ✓ ✓ ✓ ✓ ✓ ✓ ✓
Cvach 2013[33] ✓ ✓
Cvach 2014[34] ✓ ✓
Graham 2010[35] ✓
Rheineck‐Leyssius 1997[36] ✓ ✓ ✓ ✓ ✓ ✓
Taenzer 2010[31] ✓ ✓ ✓ ✓ ✓ ✓
Whalen 2014[37] ✓ ✓ ✓
Pediatric Intervention
Dandoy 2014[38] ✓ ✓ ✓ ✓ ✓ ✓

For the purposes of this review, we defined nonactionable alarms as including both invalid (false) alarms that do not that accurately represent the physiologic status of the patient and alarms that are valid but do not warrant special attention or clinical intervention (nuisance alarms). We did not separate out invalid alarms due to the tremendous variation between studies in how validity was measured.

RESULTS

Study Selection

Search results produced 4629 articles (see the flow diagram in the Supporting Information in the online version of this article), of which 32 articles were eligible: 24 observational studies describing alarm characteristics and 8 studies describing interventions to reduce alarm frequency.

Observational Study Characteristics

Characteristics of included studies are shown in Table 1. Of the 24 observational studies,[7, 8, 9, 10, 11, 12, 13, 14, 15, 16, 17, 18, 19, 20, 21, 22, 23, 24, 25, 26, 27, 28, 29, 30] 15 included adult patients,[7, 8, 9, 10, 11, 12, 13, 14, 15, 16, 17, 18, 19, 20, 21] 7 included pediatric patients,[22, 23, 24, 25, 26, 27, 28] and 2 included both adult and pediatric patients.[29, 30] All were single‐hospital studies, except for 1 study by Chambrin and colleagues[10] that included 5 sites. The number of patient‐hours examined in each study ranged from 60 to 113,880.[7, 8, 9, 10, 11, 13, 14, 15, 16, 18, 19, 20, 21, 22, 23, 24, 25, 26, 27, 29, 30] Hospital settings included ICUs (n = 16),[9, 10, 11, 13, 14, 16, 17, 18, 19, 22, 23, 24, 25, 26, 27, 29] general wards (n = 5),[12, 15, 20, 22, 28] EDs (n = 2),[7, 21] postanesthesia care unit (PACU) (n = 1),[30] and cardiac care unit (CCU) (n = 1).[8] Studies varied in the type of physiologic signals recorded and data collection methods, ranging from direct observation by a nurse who was simultaneously caring for patients[29] to video recording with expert review.[14, 19, 22] Four observational studies met the criteria for lower risk of bias.[11, 14, 15, 22]

Intervention Study Characteristics

Of the 8 intervention studies, 7 included adult patients,[31, 32, 33, 34, 35, 36, 37] and 1 included pediatric patients.[38] All were single‐hospital studies; 6 were quasi‐experimental[31, 33, 34, 35, 37, 38] and 2 were experimental.[32, 36] Settings included progressive care units (n = 3),[33, 34, 35] CCUs (n = 3),[32, 33, 37] wards (n = 2),[31, 38] PACU (n = 1),[36] and a step‐down unit (n = 1).[32] All except 1 study[32] used the monitoring system to record alarm data. Several studies evaluated multicomponent interventions that included combinations of the following: widening alarm parameters,[31, 35, 36, 37, 38] instituting alarm delays,[31, 34, 36, 38] reconfiguring alarm acuity,[35, 37] use of secondary notifications,[34] daily change of electrocardiographic electrodes or use of disposable electrocardiographic wires,[32, 33, 38] universal monitoring in high‐risk populations,[31] and timely discontinuation of monitoring in low‐risk populations.[38] Four intervention studies met our prespecified lower risk of bias criteria.[31, 32, 36, 38]

Proportion of Alarms Considered Actionable

Results of the observational studies are provided in Table 2. The proportion of alarms that were actionable was 1% to 26% in adult ICU settings,[9, 10, 11, 13, 14, 16, 17, 19] 20% to 36% in adult ward settings,[12, 15, 20] 17% in a mixed adult and pediatric PACU setting,[30] 3% to 13% in pediatric ICU settings,[22, 23, 24, 25, 26] and 1% in a pediatric ward setting.[22]

Results of Included Observational Studies
Signals Included
First Author and Publication Year Setting Monitored Patient‐Hours SpO2 ECG Arrhythmia ECG Parametersa Blood Pressure Total Alarms Actionable Alarms Alarm Response Lower Risk of Bias
  • NOTE: Lower risk of bias for observational studies required all of the following characteristics be reported: (1) two independent reviewers of alarms, (2) at least 1 clinical expert reviewer (i.e. physician or nurse), (3) the reviewer was not simultaneously involved in clinical care, and (4) there was a clear definition of alarm actionability provided in the article. Abbreviations: CCU, cardiac/telemetry care unit; ECG, electrocardiogram; ED, emergency department; ICU, intensive care unit; PACU, postanesthesia care unit; SpO2, oxygen saturation; VT, ventricular tachycardia.

  • Includes respiratory rate measured via ECG leads. Actionable is defined as alarms warranting special attention or clinical intervention. Valid is defined as the alarm accurately representing the physiologic status of the patient. Directly addresses relationship between alarm exposure and response time. ∥Not provided directly; estimated from description of data collection methods.

Adult
Atzema 2006[7] ED 371 ✓ 1,762 0.20%
Billinghurst 2003[8] CCU 420 ✓ 751 Not reported; 17% were valid Nurses with higher acuity patients and smaller % of valid alarms had slower response rates
Biot 2000[9] ICU 250 ✓ ✓ ✓ ✓ 3,665 3%
Chambrin 1999[10] ICU 1,971 ✓ ✓ ✓ ✓ 3,188 26%
Drew 2014[11] ICU 48,173 ✓ ✓ ✓ ✓ 2,558,760 0.3% of 3,861 VT alarms ✓
Gazarian 2014[12] Ward 54 nurse‐hours ✓ ✓ ✓ 205 22% Response to 47% of alarms
Grges 2009[13] ICU 200 ✓ ✓ ✓ ✓ 1,214 5%
Gross 2011[15] Ward 530 ✓ ✓ ✓ ✓ 4,393 20% ✓
Inokuchi 2013[14] ICU 2,697 ✓ ✓ ✓ ✓ 11,591 6% ✓
Koski 1990[16] ICU 400 ✓ ✓ 2,322 12%
Morales Snchez 2014[17] ICU 434 sessions ✓ ✓ ✓ 215 25% Response to 93% of alarms, of which 50% were within 10 seconds
Pergher 2014[18] ICU 60 ✓ 76 Not reported 72% of alarms stopped before nurse response or had >10 minutes response time
Siebig 2010[19] ICU 982 ✓ ✓ ✓ ✓ 5,934 15%
Voepel‐Lewis 2013[20] Ward 1,616 ✓ 710 36% Response time was longer for patients in highest quartile of total alarms
Way 2014[21] ED 93 ✓ ✓ ✓ ✓ 572 Not reported; 75% were valid Nurses responded to more alarms in resuscitation room vs acute care area, but response time was longer
Pediatric
Bonafide 2015[22] Ward + ICU 210 ✓ ✓ ✓ ✓ 5,070 13% PICU, 1% ward Incremental increases in response time as number of nonactionable alarms in preceding 120 minutes increased ✓
Lawless 1994[23] ICU 928 ✓ ✓ ✓ 2,176 6%
Rosman 2013[24] ICU 8,232 ✓ ✓ ✓ ✓ 54,656 4% of rhythm alarms true critical"
Talley 2011[25] ICU 1,470∥ ✓ ✓ ✓ ✓ 2,245 3%
Tsien 1997[26] ICU 298 ✓ ✓ ✓ 2,942 8%
van Pul 2015[27] ICU 113,880∥ ✓ ✓ ✓ ✓ 222,751 Not reported Assigned nurse did not respond to 6% of alarms within 45 seconds
Varpio 2012[28] Ward 49 unit‐hours ✓ ✓ ✓ ✓ 446 Not reported 70% of all alarms and 41% of crisis alarms were not responded to within 1 minute
Both
O'Carroll 1986[29] ICU 2,258∥ ✓ 284 2%
Wiklund 1994[30] PACU 207 ✓ ✓ ✓ 1,891 17%

Relationship Between Alarm Exposure and Response Time

Whereas 9 studies addressed response time,[8, 12, 17, 18, 20, 21, 22, 27, 28] only 2 evaluated the relationship between alarm burden and nurse response time.[20, 22] Voepel‐Lewis and colleagues found that nurse responses were slower to patients with the highest quartile of alarms (57.6 seconds) compared to those with the lowest (45.4 seconds) or medium (42.3 seconds) quartiles of alarms on an adult ward (P = 0.046). They did not find an association between false alarm exposure and response time.[20] Bonafide and colleagues found incremental increases in response time as the number of nonactionable alarms in the preceding 120 minutes increased (P 0.001 in the pediatric ICU, P = 0.009 on the pediatric ward).[22]

Interventions Effective in Reducing Alarms

Results of the 8 intervention studies are provided in Table 3. Three studies evaluated single interventions;[32, 33, 36] the remainder of the studies tested interventions with multiple components such that it was impossible to separate the effect of each component. Below, we have summarized study results, arranged by component. Because only 1 study focused on pediatric patients,[38] results from pediatric and adult settings are combined.

Results of Included Intervention Studies
First Author and Publication Year Design Setting Main Intervention Components Other/ Comments Key Results Results Statistically Significant? Lower Risk of Bias
Widen Default Settings Alarm Delays Reconfigure Alarm Acuity Secondary Notification ECG Changes
  • NOTE: Lower risk of bias for intervention studies required all of the following characteristics be reported: (1) patient census accounted for in analysis, (2) formal statistical testing of effect or statistical process control methods used to evaluate effect, (3) intervention fidelity measured and reported (defined broadly as some measurement of whether the intervention was delivered as intended), and (4) relevant patient safety outcomes measured and reported (eg, the rate of code blue events before and after the intervention was implemented). Abbreviations: CCU, cardiac/telemetry care unit; ECG, electrocardiogram; ICU, intensive care unit; ITS, interrupted time series; PACU, postanesthesia care unit; PCU, progressive care unit; SpO2, oxygen saturation. *Delays were part of secondary notification system only. Delays explored retrospectively only; not part of prospective evaluation. Preimplementation count not reported.

Adult
Albert 2015[32] Experimental (cluster‐randomized) CCU ✓ Disposable vs reusable wires Disposable leads had 29% fewer no‐telemetry, leads‐fail, and leads‐off alarms and similar artifact alarms ✓ ✓
Cvach 2013[33] Quasi‐experimental (before and after) CCU and PCU ✓ Daily change of electrodes 46% fewer alarms/bed/day
Cvach 2014[34] Quasi‐experimental (ITS) PCU ✓* ✓ Slope of regression line suggests decrease of 0.75 alarms/bed/day
Graham 2010[35] Quasi‐experimental (before and after) PCU ✓ ✓ 43% fewer crisis, warning, and system warning alarms on unit
Rheineck‐Leyssius 1997[36] Experimental (RCT) PACU ✓ ✓ Alarm limit of 85% had fewer alarms/patient but higher incidence of true hypoxemia for >1 minute (6% vs 2%) ✓ ✓
Taenzer 2010[31] Quasi‐experimental (before and after with concurrent controls) Ward ✓ ✓ Universal SpO2 monitoring Rescue events decreased from 3.4 to 1.2 per 1,000 discharges; transfers to ICU decreased from 5.6 to 2.9 per 1,000 patient‐days, only 4 alarms/patient‐day ✓ ✓
Whalen 2014[37] Quasi‐experimental (before and after) CCU ✓ ✓ 89% fewer audible alarms on unit ✓
Pediatric
Dandoy 2014[38] Quasi‐experimental (ITS) Ward ✓ ✓ ✓ Timely monitor discontinuation; daily change of ECG electrodes Decrease in alarms/patient‐days from 180 to 40 ✓

Widening alarm parameter default settings was evaluated in 5 studies:[31, 35, 36, 37, 38] 1 single intervention randomized controlled trial (RCT),[36] and 4 multiple‐intervention, quasi‐experimental studies.[31, 35, 37, 38] In the RCT, using a lower SpO2 limit of 85% instead of the standard 90% resulted in 61% fewer alarms. In the 4 multiple intervention studies, 1 study reported significant reductions in alarm rates (P 0.001),[37] 1 study did not report preintervention alarm rates but reported a postintervention alarm rate of 4 alarms per patient‐day,[31] and 2 studies reported reductions in alarm rates but did not report any statistical testing.[35, 38] Of the 3 studies examining patient safety, 1 study with universal monitoring reported fewer rescue events and transfers to the ICU postimplementation,[31] 1 study reported no missed acute decompensations,[38] and 1 study (the RCT) reported significantly more true hypoxemia events (P = 0.001).[36]

Alarm delays were evaluated in 4 studies:[31, 34, 36, 38] 3 multiple‐intervention, quasi‐experimental studies[31, 34, 38] and 1 retrospective analysis of data from an RCT.[36] One study combined alarm delays with widening defaults in a universal monitoring strategy and reported a postintervention alarm rate of 4 alarms per patient.[31] Another study evaluated delays as part of a secondary notification pager system and found a negatively sloping regression line that suggested a decreasing alarm rate, but did not report statistical testing.[34] The third study reported a reduction in alarm rates but did not report statistical testing.[38] The RCT compared the impact of a hypothetical 15‐second alarm delay to that of a lower SpO2 limit reduction and reported a similar reduction in alarms.[36] Of the 4 studies examining patient safety, 1 study with universal monitoring reported improvements,[31] 2 studies reported no adverse outcomes,[35, 38] and the retrospective analysis of data from the RCT reported the theoretical adverse outcome of delayed detection of sudden, severe desaturations.[36]

Reconfiguring alarm acuity was evaluated in 2 studies, both of which were multiple‐intervention quasi‐experimental studies.[35, 37] Both showed reductions in alarm rates: 1 was significant without increasing adverse events (P 0.001),[37] and the other did not report statistical testing or safety outcomes.[35]

Secondary notification of nurses using pagers was the main intervention component of 1 study incorporating delays between the alarms and the alarm pages.[34] As mentioned above, a negatively sloping regression line was displayed, but no statistical testing or safety outcomes were reported.

Disposable electrocardiographic lead wires or daily electrode changes were evaluated in 3 studies:[32, 33, 38] 1 single intervention cluster‐randomized trial[32] and 2 quasi‐experimental studies.[33, 38] In the cluster‐randomized trial, disposable lead wires were compared to reusable lead wires, with disposable lead wires having significantly fewer technical alarms for lead signal failures (P = 0.03) but a similar number of monitoring artifact alarms (P = 0.44).[32] In a single‐intervention, quasi‐experimental study, daily electrode change showed a reduction in alarms, but no statistical testing was reported.[33] One multiple‐intervention, quasi‐experimental study incorporating daily electrode change showed fewer alarms without statistical testing.[38] Of the 2 studies examining patient safety, both reported no adverse outcomes.[32, 38]

DISCUSSION

This systematic review of physiologic monitor alarms in the hospital yielded the following main findings: (1) between 74% and 99% of physiologic monitor alarms were not actionable, (2) a significant relationship between alarm exposure and nurse response time was demonstrated in 2 small observational studies, and (3) although interventions were most often studied in combination, results from the studies with lower risk of bias suggest that widening alarm parameters, implementing alarm delays, and using disposable electrocardiographic lead wires and/or changing electrodes daily are the most promising interventions for reducing alarms. Only 5 of 8 intervention studies measured intervention safety and found that widening alarm parameters and implementing alarm delays had mixed safety outcomes, whereas disposable electrocardiographic lead wires and daily electrode changes had no adverse safety outcomes.[29, 30, 34, 35, 36] Safety measures are crucial to ensuring the highest level of patient safety is met; interventions are rendered useless without ensuring actionable alarms are not disabled. The variation in results across studies likely reflects the wide range of care settings as well as differences in design and quality.

This field is still in its infancy, with 18 of the 32 articles published in the past 5 years. We anticipate improvements in quality and rigor as the field matures, as well as clinically tested interventions that incorporate smart alarms. Smart alarms integrate data from multiple physiologic signals and the patient's history to better detect physiologic changes in the patient and improve the positive predictive value of alarms. Academicindustry partnerships will be required to implement and rigorously test smart alarms and other emerging technologies in the hospital.

To our knowledge, this is the first systematic review focused on monitor alarms with specific review questions relevant to alarm fatigue. Cvach recently published an integrative review of alarm fatigue using research published through 2011.[39] Our review builds upon her work by contributing a more extensive and systematic search strategy with databases spanning nursing, medicine, and engineering, including additional languages, and including newer studies published through April 2015. In addition, we included multiple cross‐team checks in our eligibility review to ensure high sensitivity and specificity of the resulting set of studies.

Although we focused on interventions aiming to reduce alarms, there has also been important recent work focused on reducing telemetry utilization in adult hospital populations as well as work focused on reducing pulse oximetry utilization in children admitted with respiratory conditions. Dressler and colleagues reported an immediate and sustained reduction in telemetry utilization in hospitalized adults upon redesign of cardiac telemetry order sets to include the clinical indication, which defaulted to the American Heart Association guideline‐recommended telemetry duration.[40] Instructions for bedside nurses were also included in the order set to facilitate appropriate telemetry discontinuation. Schondelmeyer and colleagues reported reductions in continuous pulse oximetry utilization in hospitalized children with asthma and bronchiolitis upon introduction of a multifaceted quality improvement program that included provider education, a nurse handoff checklist, and discontinuation criteria incorporated into order sets.[41]

Limitations of This Review and the Underlying Body of Work

There are limitations to this systematic review and its underlying body of work. With respect to our approach to this systematic review, we focused only on monitor alarms. Numerous other medical devices generate alarms in the patient‐care environment that also can contribute to alarm fatigue and deserve equally rigorous evaluation. With respect to the underlying body of work, the quality of individual studies was generally low. For example, determinations of alarm actionability were often made by a single rater without evaluation of the reliability or validity of these determinations, and statistical testing was often missing. There were also limitations specific to intervention studies, including evaluation of nongeneralizable patient populations, failure to measure the fidelity of the interventions, inadequate measures of intervention safety, and failure to statistically evaluate alarm reductions. Finally, though not necessarily a limitation, several studies were conducted by authors involved in or funded by the medical device industry.[11, 15, 19, 31, 32] This has the potential to introduce bias, although we have no indication that the quality of the science was adversely impacted.

Moving forward, the research agenda for physiologic monitor alarms should include the following: (1) more intensive focus on evaluating the relationship between alarm exposure and response time with analysis of important mediating factors that may promote or prevent alarm fatigue, (2) emphasis on studying interventions aimed at improving alarm management using rigorous designs such as cluster‐randomized trials and trials randomized by individual participant, (3) monitoring and reporting clinically meaningful balancing measures that represent unintended consequences of disabling or delaying potentially important alarms and possibly reducing the clinicians' ability to detect true patient deterioration and intervene in a timely manner, and (4) support for transparent academicindustry partnerships to evaluate new alarm technology in real‐world settings. As evidence‐based interventions emerge, there will be new opportunities to study different implementation strategies of these interventions to optimize effectiveness.

CONCLUSIONS

The body of literature relevant to physiologic monitor alarm characteristics and alarm fatigue is limited but growing rapidly. Although we know that most alarms are not actionable and that there appears to be a relationship between alarm exposure and response time that could be caused by alarm fatigue, we cannot yet say with certainty that we know which interventions are most effective in safely reducing unnecessary alarms. Interventions that appear most promising and should be prioritized for intensive evaluation include widening alarm parameters, implementing alarm delays, and using disposable electrocardiographic lead wires and changing electrodes daily. Careful evaluation of these interventions must include systematically examining adverse patient safety consequences.

Acknowledgements

The authors thank Amogh Karnik and Micheal Sellars for their technical assistance during the review and extraction process.

Disclosures: Ms. Zander is supported by the Society of Hospital Medicine Student Hospitalist Scholar Grant. Dr. Bonafide and Ms. Stemler are supported by the National Heart, Lung, and Blood Institute of the National Institutes of Health under award number K23HL116427. The content is solely the responsibility of the authors and does not necessarily represent the official views of the National Institutes of Health. The authors report no conflicts of interest.

Files
References
  1. National Patient Safety Goals Effective January 1, 2015. The Joint Commission Web site. http://www.jointcommission.org/assets/1/6/2015_NPSG_HAP.pdf. Accessed July 17, 2015.
  2. ECRI Institute. 2015 Top 10 Health Technology Hazards. Available at: https://www.ecri.org/Pages/2015‐Hazards.aspx. Accessed June 23, 2015.
  3. Sendelbach S, Funk M. Alarm fatigue: a patient safety concern. AACN Adv Crit Care. 2013;24(4):378–386.
  4. Chopra V, McMahon LF. Redesigning hospital alarms for patient safety: alarmed and potentially dangerous. JAMA. 2014;311(12):1199–1200.
  5. Stroup DF, Berlin JA, Morton SC, et al. Meta‐analysis of observational studies in epidemiology: a proposal for reporting. Meta‐analysis Of Observational Studies in Epidemiology (MOOSE) Group. JAMA. 2000;283(15):2008–2012.
  6. Moher D, Liberati A, Tetzlaff J, Altman DG; PRISMA Group. Preferred reporting items for systematic reviews and meta‐analyses: the PRISMA statement. Ann Intern Med. 2009;151(4):264–269, W64.
  7. Atzema C, Schull MJ, Borgundvaag B, Slaughter GRD, Lee CK. ALARMED: adverse events in low‐risk patients with chest pain receiving continuous electrocardiographic monitoring in the emergency department. A pilot study. Am J Emerg Med. 2006;24:62–67.
  8. Billinghurst F, Morgan B, Arthur HM. Patient and nurse‐related implications of remote cardiac telemetry. Clin Nurs Res. 2003;12(4):356–370.
  9. Biot L, Carry PY, Perdrix JP, Eberhard A, Baconnier P. Clinical evaluation of alarm efficiency in intensive care [in French]. Ann Fr Anesth Reanim. 2000;19:459–466.
  10. Chambrin MC, Ravaux P, Calvelo‐Aros D, Jaborska A, Chopin C, Boniface B. Multicentric study of monitoring alarms in the adult intensive care unit (ICU): a descriptive analysis. Intensive Care Med. 1999;25:1360–1366.
  11. Drew BJ, Harris P, Zègre‐Hemsey JK, et al. Insights into the problem of alarm fatigue with physiologic monitor devices: a comprehensive observational study of consecutive intensive care unit patients. PloS One. 2014;9(10):e110274.
  12. Gazarian PK. Nurses' response to frequency and types of electrocardiography alarms in a non‐ critical care setting: a descriptive study. Int J Nurs Stud. 2014;51(2):190–197.
  13. Görges M, Markewitz BA, Westenskow DR. Improving alarm performance in the medical intensive care unit using delays and clinical context. Anesth Analg. 2009;108:1546–1552.
  14. Inokuchi R, Sato H, Nanjo Y, et al. The proportion of clinically relevant alarms decreases as patient clinical severity decreases in intensive care units: a pilot study. BMJ Open. 2013;3(9):e003354–e003354.
  15. Gross B, Dahl D, Nielsen L. Physiologic monitoring alarm load on medical/surgical floors of a community hospital. Biomed Instrum Technol. 2011;45:29–36.
  16. Koski EM, Mäkivirta A, Sukuvaara T, Kari A. Frequency and reliability of alarms in the monitoring of cardiac postoperative patients. Int J Clin Monit Comput. 1990;7(2):129–133.
  17. Morales Sánchez C, Murillo Pérez MA, Torrente Vela S, et al. Audit of the bedside monitor alarms in a critical care unit [in Spanish]. Enferm Intensiva. 2014;25(3):83–90.
  18. Pergher AK, Silva RCL. Stimulus‐response time to invasive blood pressure alarms: implications for the safety of critical‐care patients. Rev Gaúcha Enferm. 2014;35(2):135–141.
  19. Siebig S, Kuhls S, Imhoff M, Gather U, Scholmerich J, Wrede CE. Intensive care unit alarms— how many do we need? Crit Care Med. 2010;38:451–456.
  20. Voepel‐Lewis T, Parker ML, Burke CN, et al. Pulse oximetry desaturation alarms on a general postoperative adult unit: a prospective observational study of nurse response time. Int J Nurs Stud. 2013;50(10):1351–1358.
  21. Way RB, Beer SA, Wilson SJ. Whats that noise? Bedside monitoring in the Emergency Department. Int Emerg Nurs. 2014;22(4):197–201.
  22. Bonafide CP, Lin R, Zander M, et al. Association between exposure to nonactionable physiologic monitor alarms and response time in a children's hospital. J Hosp Med. 2015;10(6):345–351.
  23. Lawless ST. Crying wolf: false alarms in a pediatric intensive care unit. Crit Care Med. 1994;22(6):981–985.
  24. Rosman EC, Blaufox AD, Menco A, Trope R, Seiden HS. What are we missing? Arrhythmia detection in the pediatric intensive care unit. J Pediatr. 2013;163(2):511–514.
  25. Talley LB, Hooper J, Jacobs B, et al. Cardiopulmonary monitors and clinically significant events in critically ill children. Biomed Instrum Technol. 2011;45(s1):38–45.
  26. Tsien CL, Fackler JC. Poor prognosis for existing monitors in the intensive care unit. Crit Care Med. 1997;25:614–619.
  27. Pul C, Mortel H, Bogaart J, Mohns T, Andriessen P. Safe patient monitoring is challenging but still feasible in a neonatal intensive care unit with single family rooms. Acta Paediatr Oslo Nor 1992. 2015;104(6):e247–e254.
  28. Varpio L, Kuziemsky C, Macdonald C, King WJ. The helpful or hindering effects of in‐hospital patient monitor alarms on nurses: a qualitative analysis. CIN Comput Inform Nurs. 2012;30(4):210–217.
  29. O'Carroll T. Survey of alarms in an intensive therapy unit. Anaesthesia. 1986;41(7):742–744.
  30. Wiklund L, Hök B, Ståhl K, Jordeby‐Jönsson A. Postanesthesia monitoring revisited: frequency of true and false alarms from different monitoring devices. J Clin Anesth. 1994;6(3):182–188.
  31. Taenzer AH, Pyke JB, McGrath SP, Blike GT. Impact of pulse oximetry surveillance on rescue events and intensive care unit transfers: a before‐and‐after concurrence study. Anesthesiology. 2010;112(2):282–287.
  32. Albert NM, Murray T, Bena JF, et al. Differences in alarm events between disposable and reusable electrocardiography lead wires. Am J Crit Care. 2015;24(1):67–74.
  33. Cvach MM, Biggs M, Rothwell KJ, Charles‐Hudson C. Daily electrode change and effect on cardiac monitor alarms: an evidence‐based practice approach. J Nurs Care Qual. 2013;28:265–271.
  34. Cvach MM, Frank RJ, Doyle P, Stevens ZK. Use of pagers with an alarm escalation system to reduce cardiac monitor alarm signals. J Nurs Care Qual. 2014;29(1):9–18.
  35. Graham KC, Cvach M. Monitor alarm fatigue: standardizing use of physiological monitors and decreasing nuisance alarms. Am J Crit Care. 2010;19:28–34.
  36. Rheineck‐Leyssius AT, Kalkman CJ. Influence of pulse oximeter lower alarm limit on the incidence of hypoxaemia in the recovery room. Br J Anaesth. 1997;79(4):460–464.
  37. Whalen DA, Covelle PM, Piepenbrink JC, Villanova KL, Cuneo CL, Awtry EH. Novel approach to cardiac alarm management on telemetry units. J Cardiovasc Nurs. 2014;29(5):E13–E22.
  38. Dandoy CE, Davies SM, Flesch L, et al. A team‐based approach to reducing cardiac monitor alarms. Pediatrics. 2014;134(6):e1686–e1694.
  39. Cvach M. Monitor alarm fatigue: an integrative review. Biomed Instrum Technol. 2012;46(4):268–277.
  40. Dressler R, Dryer MM, Coletti C, Mahoney D, Doorey AJ. Altering overuse of cardiac telemetry in non‐intensive care unit settings by hardwiring the use of American Heart Association guidelines. JAMA Intern Med. 2014;174(11):1852–1854.
  41. Schondelmeyer AC, Simmons JM, Statile AM, et al. Using quality improvement to reduce continuous pulse oximetry use in children with wheezing. Pediatrics. 2015;135(4):e1044–e1051.
  42. Downs SH, Black N. The feasibility of creating a checklist for the assessment of the methodological quality both of randomised and non‐randomised studies of health care interventions. J Epidemiol Community Health. 1998;52(6):377–384.
Article PDF
Issue
Journal of Hospital Medicine - 11(2)
Page Number
136-144
Sections
Files
Files
Article PDF
Article PDF

Clinical alarm safety has become a recent target for improvement in many hospitals. In 2013, The Joint Commission released a National Patient Safety Goal prompting accredited hospitals to establish alarm safety as a hospital priority, identify the most important alarm signals to manage, and, by 2016, develop policies and procedures that address alarm management.[1] In addition, the Emergency Care Research Institute has named alarm hazards the top health technology hazard each year since 2012.[2]

The primary arguments supporting the elevation of alarm management to a national hospital priority in the United States include the following: (1) clinicians rely on alarms to notify them of important physiologic changes, (2) alarms occur frequently and usually do not warrant clinical intervention, and (3) alarm overload renders clinicians unable to respond to all alarms, resulting in alarm fatigue: responding more slowly or ignoring alarms that may represent actual clinical deterioration.[3, 4] These arguments are built largely on anecdotal data, reported safety event databases, and small studies that have not previously been systematically analyzed.

Despite the national focus on alarms, we still know very little about fundamental questions key to improving alarm safety. In this systematic review, we aimed to answer 3 key questions about physiologic monitor alarms: (1) What proportion of alarms warrant attention or clinical intervention (ie, actionable alarms), and how does this proportion vary between adult and pediatric populations and between intensive care unit (ICU) and ward settings? (2) What is the relationship between alarm exposure and clinician response time? (3) What interventions are effective in reducing the frequency of alarms?

We limited our scope to monitor alarms because few studies have evaluated the characteristics of alarms from other medical devices, and because missing relevant monitor alarms could adversely impact patient safety.

METHODS

We performed a systematic review of the literature in accordance with the Meta‐Analysis of Observational Studies in Epidemiology guidelines[5] and developed this manuscript using the Preferred Reporting Items for Systematic Reviews and Meta‐Analyses (PRISMA) statement.[6]

Eligibility Criteria

With help from an experienced biomedical librarian (C.D.S.), we searched PubMed, the Cumulative Index to Nursing and Allied Health Literature, Scopus, Cochrane Library, ClinicalTrials.gov, and Google Scholar from January 1980 through April 2015 (see Supporting Information in the online version of this article for the search terms and queries). We hand searched the reference lists of included articles and reviewed our personal libraries to identify additional relevant studies.

We included peer‐reviewed, original research studies published in English, Spanish, or French that addressed the questions outlined above. Eligible patient populations were children and adults admitted to hospital inpatient units and emergency departments (EDs). We excluded alarms in procedural suites or operating rooms (typically responded to by anesthesiologists already with the patient) because of the differences in environment of care, staff‐to‐patient ratio, and equipment. We included observational studies reporting the actionability of physiologic monitor alarms (ie, alarms warranting special attention or clinical intervention), as well as nurse responses to these alarms. We excluded studies focused on the effects of alarms unrelated to patient safety, such as families' and patients' stress, noise, or sleep disturbance. We included only intervention studies evaluating pragmatic interventions ready for clinical implementation (ie, not experimental devices or software algorithms).

Selection Process and Data Extraction

First, 2 authors screened the titles and abstracts of articles for eligibility. To maximize sensitivity, if at least 1 author considered the article relevant, the article proceeded to full‐text review. Second, the full texts of articles screened were independently reviewed by 2 authors in an unblinded fashion to determine their eligibility. Any disagreements concerning eligibility were resolved by team consensus. To assure consistency in eligibility determinations across the team, a core group of the authors (C.W.P, C.P.B., E.E., and V.V.G.) held a series of meetings to review and discuss each potentially eligible article and reach consensus on the final list of included articles. Two authors independently extracted the following characteristics from included studies: alarm review methods, analytic design, fidelity measurement, consideration of unintended adverse safety consequences, and key results. Reviewers were not blinded to journal, authors, or affiliations.

Synthesis of Results and Risk Assessment

Given the high degree of heterogeneity in methodology, we were unable to generate summary proportions of the observational studies or perform a meta‐analysis of the intervention studies. Thus, we organized the studies into clinically relevant categories and presented key aspects in tables. Due to the heterogeneity of the studies and the controversy surrounding quality scores,[5] we did not generate summary scores of study quality. Instead, we evaluated and reported key design elements that had the potential to bias the results. To recognize the more comprehensive studies in the field, we developed by consensus a set of characteristics that distinguished studies with lower risk of bias. These characteristics are shown and defined in Table 1.

General Characteristics of Included Studies
First Author and Publication Year Alarm Review Method Indicators of Potential Bias for Observational Studies Indicators of Potential Bias for Intervention Studies
Monitor System Direct Observation Medical Record Review Rhythm Annotation Video Observation Remote Monitoring Staff Medical Device Industry Involved Two Independent Reviewers At Least 1 Reviewer Is a Clinical Expert Reviewer Not Simultaneously in Patient Care Clear Definition of Alarm Actionability Census Included Statistical Testing or QI SPC Methods Fidelity Assessed Safety Assessed Lower Risk of Bias
  • NOTE: Lower risk of bias for observational studies required all of the following characteristics be reported: (1) two independent reviewers of alarms, (2) at least 1 clinical expert reviewer (ie, physician or nurse), (3) the reviewer was not simultaneously involved in clinical care, and (4) there was a clear definition of alarm actionability provided in the article. These indicators assess detection bias, observer bias, analytical bias, and reporting bias and were derived from the Meta‐analysis of Observational Studies in Epidemiology checklist.[5] Lower risk of bias for intervention studies required all of the following characteristics be reported: (1) patient census accounted for in analysis, (2) formal statistical testing of effect or statistical process control methods used to evaluate effect, (3) intervention fidelity measured and reported (defined broadly as some measurement of whether the intervention was delivered as intended), and (4) relevant patient safety outcomes measured and reported (eg, the rate of code blue events before and after the intervention was implemented). These indicators assess reporting bias and internal validity bias and were derived from the Downs and Black checklist.[42] Monitor system: alarm data were electronically collected directly from the physiologic monitors and saved on a computer device through software such as BedMasterEx. Direct observation: an in‐person observer, such as a research assistant or a nurse, takes note of the alarm data and/or responses to alarms. Medical record review: data on alarms and/or responses to alarms were extracted from the patient medical records. Rhythm annotation: data on waveforms from cardiac monitors were collected and saved on a computer device through software such as BedMasterEx. Video observation: video cameras were set up in the patient's room and recorded data on alarms and/or responses to alarms. Remote monitor staff: clinicians situated at a remote location observe the patient via video camera and may be able to communicate with the patient or the patient's assigned nurse. Abbreviations: QI, quality improvement; RN, registered nurse; SPC, statistical process control. *Monitor system + RN interrogation. Assigned nurse making observations. Monitor from central station. Alarm outcome reported using run chart, and fidelity outcomes presented using statistical process control charts.

Adult Observational
Atzema 2006[7] ✓* ✓ ✓
Billinghurst 2003[8] ✓ ✓ ✓ ✓
Biot 2000[9] ✓ ✓ ✓ ✓
Chambrin 1999[10] ✓ ✓ ✓ ✓
Drew 2014[11] ✓ ✓ ✓ ✓ ✓ ✓ ✓ ✓
Gazarian 2014[12] ✓ ✓ ✓ ✓ ✓
Grges 2009[13] ✓ ✓ ✓ ✓
Gross 2011[15] ✓ ✓ ✓ ✓ ✓ ✓ ✓
Inokuchi 2013[14] ✓ ✓ ✓ ✓ ✓ ✓ ✓
Koski 1990[16] ✓ ✓ ✓ ✓
Morales Snchez 2014[17] ✓ ✓ ✓ ✓
Pergher 2014[18] ✓ ✓
Siebig 2010[19] ✓ ✓ ✓ ✓ ✓ ✓
Voepel‐Lewis 2013[20] ✓ ✓ ✓ ✓
Way 2014[21] ✓ ✓ ✓
Pediatric Observational
Bonafide 2015[22] ✓ ✓ ✓ ✓ ✓ ✓ ✓
Lawless 1994[23] ✓ ✓
Rosman 2013[24] ✓ ✓ ✓ ✓ ✓
Talley 2011[25] ✓ ✓ ✓ ✓ ✓
Tsien 1997[26] ✓ ✓ ✓ ✓
van Pul 2015[27] ✓
Varpio 2012[28] ✓ ✓ ✓ ✓
Mixed Adult and Pediatric Observational
O'Carroll 1986[29] ✓
Wiklund 1994[30] ✓ ✓ ✓ ✓
Adult Intervention
Albert 2015[32] ✓ ✓ ✓ ✓ ✓ ✓ ✓
Cvach 2013[33] ✓ ✓
Cvach 2014[34] ✓ ✓
Graham 2010[35] ✓
Rheineck‐Leyssius 1997[36] ✓ ✓ ✓ ✓ ✓ ✓
Taenzer 2010[31] ✓ ✓ ✓ ✓ ✓ ✓
Whalen 2014[37] ✓ ✓ ✓
Pediatric Intervention
Dandoy 2014[38] ✓ ✓ ✓ ✓ ✓ ✓

For the purposes of this review, we defined nonactionable alarms as including both invalid (false) alarms that do not that accurately represent the physiologic status of the patient and alarms that are valid but do not warrant special attention or clinical intervention (nuisance alarms). We did not separate out invalid alarms due to the tremendous variation between studies in how validity was measured.

RESULTS

Study Selection

Search results produced 4629 articles (see the flow diagram in the Supporting Information in the online version of this article), of which 32 articles were eligible: 24 observational studies describing alarm characteristics and 8 studies describing interventions to reduce alarm frequency.

Observational Study Characteristics

Characteristics of included studies are shown in Table 1. Of the 24 observational studies,[7, 8, 9, 10, 11, 12, 13, 14, 15, 16, 17, 18, 19, 20, 21, 22, 23, 24, 25, 26, 27, 28, 29, 30] 15 included adult patients,[7, 8, 9, 10, 11, 12, 13, 14, 15, 16, 17, 18, 19, 20, 21] 7 included pediatric patients,[22, 23, 24, 25, 26, 27, 28] and 2 included both adult and pediatric patients.[29, 30] All were single‐hospital studies, except for 1 study by Chambrin and colleagues[10] that included 5 sites. The number of patient‐hours examined in each study ranged from 60 to 113,880.[7, 8, 9, 10, 11, 13, 14, 15, 16, 18, 19, 20, 21, 22, 23, 24, 25, 26, 27, 29, 30] Hospital settings included ICUs (n = 16),[9, 10, 11, 13, 14, 16, 17, 18, 19, 22, 23, 24, 25, 26, 27, 29] general wards (n = 5),[12, 15, 20, 22, 28] EDs (n = 2),[7, 21] postanesthesia care unit (PACU) (n = 1),[30] and cardiac care unit (CCU) (n = 1).[8] Studies varied in the type of physiologic signals recorded and data collection methods, ranging from direct observation by a nurse who was simultaneously caring for patients[29] to video recording with expert review.[14, 19, 22] Four observational studies met the criteria for lower risk of bias.[11, 14, 15, 22]

Intervention Study Characteristics

Of the 8 intervention studies, 7 included adult patients,[31, 32, 33, 34, 35, 36, 37] and 1 included pediatric patients.[38] All were single‐hospital studies; 6 were quasi‐experimental[31, 33, 34, 35, 37, 38] and 2 were experimental.[32, 36] Settings included progressive care units (n = 3),[33, 34, 35] CCUs (n = 3),[32, 33, 37] wards (n = 2),[31, 38] PACU (n = 1),[36] and a step‐down unit (n = 1).[32] All except 1 study[32] used the monitoring system to record alarm data. Several studies evaluated multicomponent interventions that included combinations of the following: widening alarm parameters,[31, 35, 36, 37, 38] instituting alarm delays,[31, 34, 36, 38] reconfiguring alarm acuity,[35, 37] use of secondary notifications,[34] daily change of electrocardiographic electrodes or use of disposable electrocardiographic wires,[32, 33, 38] universal monitoring in high‐risk populations,[31] and timely discontinuation of monitoring in low‐risk populations.[38] Four intervention studies met our prespecified lower risk of bias criteria.[31, 32, 36, 38]

Proportion of Alarms Considered Actionable

Results of the observational studies are provided in Table 2. The proportion of alarms that were actionable was 1% to 26% in adult ICU settings,[9, 10, 11, 13, 14, 16, 17, 19] 20% to 36% in adult ward settings,[12, 15, 20] 17% in a mixed adult and pediatric PACU setting,[30] 3% to 13% in pediatric ICU settings,[22, 23, 24, 25, 26] and 1% in a pediatric ward setting.[22]

Results of Included Observational Studies
Signals Included
First Author and Publication Year Setting Monitored Patient‐Hours SpO2 ECG Arrhythmia ECG Parametersa Blood Pressure Total Alarms Actionable Alarms Alarm Response Lower Risk of Bias
  • NOTE: Lower risk of bias for observational studies required all of the following characteristics be reported: (1) two independent reviewers of alarms, (2) at least 1 clinical expert reviewer (i.e. physician or nurse), (3) the reviewer was not simultaneously involved in clinical care, and (4) there was a clear definition of alarm actionability provided in the article. Abbreviations: CCU, cardiac/telemetry care unit; ECG, electrocardiogram; ED, emergency department; ICU, intensive care unit; PACU, postanesthesia care unit; SpO2, oxygen saturation; VT, ventricular tachycardia.

  • Includes respiratory rate measured via ECG leads. Actionable is defined as alarms warranting special attention or clinical intervention. Valid is defined as the alarm accurately representing the physiologic status of the patient. Directly addresses relationship between alarm exposure and response time. ∥Not provided directly; estimated from description of data collection methods.

Adult
Atzema 2006[7] ED 371 ✓ 1,762 0.20%
Billinghurst 2003[8] CCU 420 ✓ 751 Not reported; 17% were valid Nurses with higher acuity patients and smaller % of valid alarms had slower response rates
Biot 2000[9] ICU 250 ✓ ✓ ✓ ✓ 3,665 3%
Chambrin 1999[10] ICU 1,971 ✓ ✓ ✓ ✓ 3,188 26%
Drew 2014[11] ICU 48,173 ✓ ✓ ✓ ✓ 2,558,760 0.3% of 3,861 VT alarms ✓
Gazarian 2014[12] Ward 54 nurse‐hours ✓ ✓ ✓ 205 22% Response to 47% of alarms
Grges 2009[13] ICU 200 ✓ ✓ ✓ ✓ 1,214 5%
Gross 2011[15] Ward 530 ✓ ✓ ✓ ✓ 4,393 20% ✓
Inokuchi 2013[14] ICU 2,697 ✓ ✓ ✓ ✓ 11,591 6% ✓
Koski 1990[16] ICU 400 ✓ ✓ 2,322 12%
Morales Snchez 2014[17] ICU 434 sessions ✓ ✓ ✓ 215 25% Response to 93% of alarms, of which 50% were within 10 seconds
Pergher 2014[18] ICU 60 ✓ 76 Not reported 72% of alarms stopped before nurse response or had >10 minutes response time
Siebig 2010[19] ICU 982 ✓ ✓ ✓ ✓ 5,934 15%
Voepel‐Lewis 2013[20] Ward 1,616 ✓ 710 36% Response time was longer for patients in highest quartile of total alarms
Way 2014[21] ED 93 ✓ ✓ ✓ ✓ 572 Not reported; 75% were valid Nurses responded to more alarms in resuscitation room vs acute care area, but response time was longer
Pediatric
Bonafide 2015[22] Ward + ICU 210 ✓ ✓ ✓ ✓ 5,070 13% PICU, 1% ward Incremental increases in response time as number of nonactionable alarms in preceding 120 minutes increased ✓
Lawless 1994[23] ICU 928 ✓ ✓ ✓ 2,176 6%
Rosman 2013[24] ICU 8,232 ✓ ✓ ✓ ✓ 54,656 4% of rhythm alarms true critical"
Talley 2011[25] ICU 1,470∥ ✓ ✓ ✓ ✓ 2,245 3%
Tsien 1997[26] ICU 298 ✓ ✓ ✓ 2,942 8%
van Pul 2015[27] ICU 113,880∥ ✓ ✓ ✓ ✓ 222,751 Not reported Assigned nurse did not respond to 6% of alarms within 45 seconds
Varpio 2012[28] Ward 49 unit‐hours ✓ ✓ ✓ ✓ 446 Not reported 70% of all alarms and 41% of crisis alarms were not responded to within 1 minute
Both
O'Carroll 1986[29] ICU 2,258∥ ✓ 284 2%
Wiklund 1994[30] PACU 207 ✓ ✓ ✓ 1,891 17%

Relationship Between Alarm Exposure and Response Time

Whereas 9 studies addressed response time,[8, 12, 17, 18, 20, 21, 22, 27, 28] only 2 evaluated the relationship between alarm burden and nurse response time.[20, 22] Voepel‐Lewis and colleagues found that nurse responses were slower to patients with the highest quartile of alarms (57.6 seconds) compared to those with the lowest (45.4 seconds) or medium (42.3 seconds) quartiles of alarms on an adult ward (P = 0.046). They did not find an association between false alarm exposure and response time.[20] Bonafide and colleagues found incremental increases in response time as the number of nonactionable alarms in the preceding 120 minutes increased (P 0.001 in the pediatric ICU, P = 0.009 on the pediatric ward).[22]

Interventions Effective in Reducing Alarms

Results of the 8 intervention studies are provided in Table 3. Three studies evaluated single interventions;[32, 33, 36] the remainder of the studies tested interventions with multiple components such that it was impossible to separate the effect of each component. Below, we have summarized study results, arranged by component. Because only 1 study focused on pediatric patients,[38] results from pediatric and adult settings are combined.

Results of Included Intervention Studies
First Author and Publication Year Design Setting Main Intervention Components Other/ Comments Key Results Results Statistically Significant? Lower Risk of Bias
Widen Default Settings Alarm Delays Reconfigure Alarm Acuity Secondary Notification ECG Changes
  • NOTE: Lower risk of bias for intervention studies required all of the following characteristics be reported: (1) patient census accounted for in analysis, (2) formal statistical testing of effect or statistical process control methods used to evaluate effect, (3) intervention fidelity measured and reported (defined broadly as some measurement of whether the intervention was delivered as intended), and (4) relevant patient safety outcomes measured and reported (eg, the rate of code blue events before and after the intervention was implemented). Abbreviations: CCU, cardiac/telemetry care unit; ECG, electrocardiogram; ICU, intensive care unit; ITS, interrupted time series; PACU, postanesthesia care unit; PCU, progressive care unit; SpO2, oxygen saturation. *Delays were part of secondary notification system only. Delays explored retrospectively only; not part of prospective evaluation. Preimplementation count not reported.

Adult
Albert 2015[32] Experimental (cluster‐randomized) CCU ✓ Disposable vs reusable wires Disposable leads had 29% fewer no‐telemetry, leads‐fail, and leads‐off alarms and similar artifact alarms ✓ ✓
Cvach 2013[33] Quasi‐experimental (before and after) CCU and PCU ✓ Daily change of electrodes 46% fewer alarms/bed/day
Cvach 2014[34] Quasi‐experimental (ITS) PCU ✓* ✓ Slope of regression line suggests decrease of 0.75 alarms/bed/day
Graham 2010[35] Quasi‐experimental (before and after) PCU ✓ ✓ 43% fewer crisis, warning, and system warning alarms on unit
Rheineck‐Leyssius 1997[36] Experimental (RCT) PACU ✓ ✓ Alarm limit of 85% had fewer alarms/patient but higher incidence of true hypoxemia for >1 minute (6% vs 2%) ✓ ✓
Taenzer 2010[31] Quasi‐experimental (before and after with concurrent controls) Ward ✓ ✓ Universal SpO2 monitoring Rescue events decreased from 3.4 to 1.2 per 1,000 discharges; transfers to ICU decreased from 5.6 to 2.9 per 1,000 patient‐days, only 4 alarms/patient‐day ✓ ✓
Whalen 2014[37] Quasi‐experimental (before and after) CCU ✓ ✓ 89% fewer audible alarms on unit ✓
Pediatric
Dandoy 2014[38] Quasi‐experimental (ITS) Ward ✓ ✓ ✓ Timely monitor discontinuation; daily change of ECG electrodes Decrease in alarms/patient‐days from 180 to 40 ✓

Widening alarm parameter default settings was evaluated in 5 studies:[31, 35, 36, 37, 38] 1 single intervention randomized controlled trial (RCT),[36] and 4 multiple‐intervention, quasi‐experimental studies.[31, 35, 37, 38] In the RCT, using a lower SpO2 limit of 85% instead of the standard 90% resulted in 61% fewer alarms. In the 4 multiple intervention studies, 1 study reported significant reductions in alarm rates (P 0.001),[37] 1 study did not report preintervention alarm rates but reported a postintervention alarm rate of 4 alarms per patient‐day,[31] and 2 studies reported reductions in alarm rates but did not report any statistical testing.[35, 38] Of the 3 studies examining patient safety, 1 study with universal monitoring reported fewer rescue events and transfers to the ICU postimplementation,[31] 1 study reported no missed acute decompensations,[38] and 1 study (the RCT) reported significantly more true hypoxemia events (P = 0.001).[36]

Alarm delays were evaluated in 4 studies:[31, 34, 36, 38] 3 multiple‐intervention, quasi‐experimental studies[31, 34, 38] and 1 retrospective analysis of data from an RCT.[36] One study combined alarm delays with widening defaults in a universal monitoring strategy and reported a postintervention alarm rate of 4 alarms per patient.[31] Another study evaluated delays as part of a secondary notification pager system and found a negatively sloping regression line that suggested a decreasing alarm rate, but did not report statistical testing.[34] The third study reported a reduction in alarm rates but did not report statistical testing.[38] The RCT compared the impact of a hypothetical 15‐second alarm delay to that of a lower SpO2 limit reduction and reported a similar reduction in alarms.[36] Of the 4 studies examining patient safety, 1 study with universal monitoring reported improvements,[31] 2 studies reported no adverse outcomes,[35, 38] and the retrospective analysis of data from the RCT reported the theoretical adverse outcome of delayed detection of sudden, severe desaturations.[36]

Reconfiguring alarm acuity was evaluated in 2 studies, both of which were multiple‐intervention quasi‐experimental studies.[35, 37] Both showed reductions in alarm rates: 1 was significant without increasing adverse events (P 0.001),[37] and the other did not report statistical testing or safety outcomes.[35]

Secondary notification of nurses using pagers was the main intervention component of 1 study incorporating delays between the alarms and the alarm pages.[34] As mentioned above, a negatively sloping regression line was displayed, but no statistical testing or safety outcomes were reported.

Disposable electrocardiographic lead wires or daily electrode changes were evaluated in 3 studies:[32, 33, 38] 1 single intervention cluster‐randomized trial[32] and 2 quasi‐experimental studies.[33, 38] In the cluster‐randomized trial, disposable lead wires were compared to reusable lead wires, with disposable lead wires having significantly fewer technical alarms for lead signal failures (P = 0.03) but a similar number of monitoring artifact alarms (P = 0.44).[32] In a single‐intervention, quasi‐experimental study, daily electrode change showed a reduction in alarms, but no statistical testing was reported.[33] One multiple‐intervention, quasi‐experimental study incorporating daily electrode change showed fewer alarms without statistical testing.[38] Of the 2 studies examining patient safety, both reported no adverse outcomes.[32, 38]

DISCUSSION

This systematic review of physiologic monitor alarms in the hospital yielded the following main findings: (1) between 74% and 99% of physiologic monitor alarms were not actionable, (2) a significant relationship between alarm exposure and nurse response time was demonstrated in 2 small observational studies, and (3) although interventions were most often studied in combination, results from the studies with lower risk of bias suggest that widening alarm parameters, implementing alarm delays, and using disposable electrocardiographic lead wires and/or changing electrodes daily are the most promising interventions for reducing alarms. Only 5 of 8 intervention studies measured intervention safety and found that widening alarm parameters and implementing alarm delays had mixed safety outcomes, whereas disposable electrocardiographic lead wires and daily electrode changes had no adverse safety outcomes.[29, 30, 34, 35, 36] Safety measures are crucial to ensuring the highest level of patient safety is met; interventions are rendered useless without ensuring actionable alarms are not disabled. The variation in results across studies likely reflects the wide range of care settings as well as differences in design and quality.

This field is still in its infancy, with 18 of the 32 articles published in the past 5 years. We anticipate improvements in quality and rigor as the field matures, as well as clinically tested interventions that incorporate smart alarms. Smart alarms integrate data from multiple physiologic signals and the patient's history to better detect physiologic changes in the patient and improve the positive predictive value of alarms. Academicindustry partnerships will be required to implement and rigorously test smart alarms and other emerging technologies in the hospital.

To our knowledge, this is the first systematic review focused on monitor alarms with specific review questions relevant to alarm fatigue. Cvach recently published an integrative review of alarm fatigue using research published through 2011.[39] Our review builds upon her work by contributing a more extensive and systematic search strategy with databases spanning nursing, medicine, and engineering, including additional languages, and including newer studies published through April 2015. In addition, we included multiple cross‐team checks in our eligibility review to ensure high sensitivity and specificity of the resulting set of studies.

Although we focused on interventions aiming to reduce alarms, there has also been important recent work focused on reducing telemetry utilization in adult hospital populations as well as work focused on reducing pulse oximetry utilization in children admitted with respiratory conditions. Dressler and colleagues reported an immediate and sustained reduction in telemetry utilization in hospitalized adults upon redesign of cardiac telemetry order sets to include the clinical indication, which defaulted to the American Heart Association guideline‐recommended telemetry duration.[40] Instructions for bedside nurses were also included in the order set to facilitate appropriate telemetry discontinuation. Schondelmeyer and colleagues reported reductions in continuous pulse oximetry utilization in hospitalized children with asthma and bronchiolitis upon introduction of a multifaceted quality improvement program that included provider education, a nurse handoff checklist, and discontinuation criteria incorporated into order sets.[41]

Limitations of This Review and the Underlying Body of Work

There are limitations to this systematic review and its underlying body of work. With respect to our approach to this systematic review, we focused only on monitor alarms. Numerous other medical devices generate alarms in the patient‐care environment that also can contribute to alarm fatigue and deserve equally rigorous evaluation. With respect to the underlying body of work, the quality of individual studies was generally low. For example, determinations of alarm actionability were often made by a single rater without evaluation of the reliability or validity of these determinations, and statistical testing was often missing. There were also limitations specific to intervention studies, including evaluation of nongeneralizable patient populations, failure to measure the fidelity of the interventions, inadequate measures of intervention safety, and failure to statistically evaluate alarm reductions. Finally, though not necessarily a limitation, several studies were conducted by authors involved in or funded by the medical device industry.[11, 15, 19, 31, 32] This has the potential to introduce bias, although we have no indication that the quality of the science was adversely impacted.

Moving forward, the research agenda for physiologic monitor alarms should include the following: (1) more intensive focus on evaluating the relationship between alarm exposure and response time with analysis of important mediating factors that may promote or prevent alarm fatigue, (2) emphasis on studying interventions aimed at improving alarm management using rigorous designs such as cluster‐randomized trials and trials randomized by individual participant, (3) monitoring and reporting clinically meaningful balancing measures that represent unintended consequences of disabling or delaying potentially important alarms and possibly reducing the clinicians' ability to detect true patient deterioration and intervene in a timely manner, and (4) support for transparent academicindustry partnerships to evaluate new alarm technology in real‐world settings. As evidence‐based interventions emerge, there will be new opportunities to study different implementation strategies of these interventions to optimize effectiveness.

CONCLUSIONS

The body of literature relevant to physiologic monitor alarm characteristics and alarm fatigue is limited but growing rapidly. Although we know that most alarms are not actionable and that there appears to be a relationship between alarm exposure and response time that could be caused by alarm fatigue, we cannot yet say with certainty that we know which interventions are most effective in safely reducing unnecessary alarms. Interventions that appear most promising and should be prioritized for intensive evaluation include widening alarm parameters, implementing alarm delays, and using disposable electrocardiographic lead wires and changing electrodes daily. Careful evaluation of these interventions must include systematically examining adverse patient safety consequences.

Acknowledgements

The authors thank Amogh Karnik and Micheal Sellars for their technical assistance during the review and extraction process.

Disclosures: Ms. Zander is supported by the Society of Hospital Medicine Student Hospitalist Scholar Grant. Dr. Bonafide and Ms. Stemler are supported by the National Heart, Lung, and Blood Institute of the National Institutes of Health under award number K23HL116427. The content is solely the responsibility of the authors and does not necessarily represent the official views of the National Institutes of Health. The authors report no conflicts of interest.

Clinical alarm safety has become a recent target for improvement in many hospitals. In 2013, The Joint Commission released a National Patient Safety Goal prompting accredited hospitals to establish alarm safety as a hospital priority, identify the most important alarm signals to manage, and, by 2016, develop policies and procedures that address alarm management.[1] In addition, the Emergency Care Research Institute has named alarm hazards the top health technology hazard each year since 2012.[2]

The primary arguments supporting the elevation of alarm management to a national hospital priority in the United States include the following: (1) clinicians rely on alarms to notify them of important physiologic changes, (2) alarms occur frequently and usually do not warrant clinical intervention, and (3) alarm overload renders clinicians unable to respond to all alarms, resulting in alarm fatigue: responding more slowly or ignoring alarms that may represent actual clinical deterioration.[3, 4] These arguments are built largely on anecdotal data, reported safety event databases, and small studies that have not previously been systematically analyzed.

Despite the national focus on alarms, we still know very little about fundamental questions key to improving alarm safety. In this systematic review, we aimed to answer 3 key questions about physiologic monitor alarms: (1) What proportion of alarms warrant attention or clinical intervention (ie, actionable alarms), and how does this proportion vary between adult and pediatric populations and between intensive care unit (ICU) and ward settings? (2) What is the relationship between alarm exposure and clinician response time? (3) What interventions are effective in reducing the frequency of alarms?

We limited our scope to monitor alarms because few studies have evaluated the characteristics of alarms from other medical devices, and because missing relevant monitor alarms could adversely impact patient safety.

METHODS

We performed a systematic review of the literature in accordance with the Meta‐Analysis of Observational Studies in Epidemiology guidelines[5] and developed this manuscript using the Preferred Reporting Items for Systematic Reviews and Meta‐Analyses (PRISMA) statement.[6]

Eligibility Criteria

With help from an experienced biomedical librarian (C.D.S.), we searched PubMed, the Cumulative Index to Nursing and Allied Health Literature, Scopus, Cochrane Library, ClinicalTrials.gov, and Google Scholar from January 1980 through April 2015 (see Supporting Information in the online version of this article for the search terms and queries). We hand searched the reference lists of included articles and reviewed our personal libraries to identify additional relevant studies.

We included peer‐reviewed, original research studies published in English, Spanish, or French that addressed the questions outlined above. Eligible patient populations were children and adults admitted to hospital inpatient units and emergency departments (EDs). We excluded alarms in procedural suites or operating rooms (typically responded to by anesthesiologists already with the patient) because of the differences in environment of care, staff‐to‐patient ratio, and equipment. We included observational studies reporting the actionability of physiologic monitor alarms (ie, alarms warranting special attention or clinical intervention), as well as nurse responses to these alarms. We excluded studies focused on the effects of alarms unrelated to patient safety, such as families' and patients' stress, noise, or sleep disturbance. We included only intervention studies evaluating pragmatic interventions ready for clinical implementation (ie, not experimental devices or software algorithms).

Selection Process and Data Extraction

First, 2 authors screened the titles and abstracts of articles for eligibility. To maximize sensitivity, if at least 1 author considered the article relevant, the article proceeded to full‐text review. Second, the full texts of articles screened were independently reviewed by 2 authors in an unblinded fashion to determine their eligibility. Any disagreements concerning eligibility were resolved by team consensus. To assure consistency in eligibility determinations across the team, a core group of the authors (C.W.P, C.P.B., E.E., and V.V.G.) held a series of meetings to review and discuss each potentially eligible article and reach consensus on the final list of included articles. Two authors independently extracted the following characteristics from included studies: alarm review methods, analytic design, fidelity measurement, consideration of unintended adverse safety consequences, and key results. Reviewers were not blinded to journal, authors, or affiliations.

Synthesis of Results and Risk Assessment

Given the high degree of heterogeneity in methodology, we were unable to generate summary proportions of the observational studies or perform a meta‐analysis of the intervention studies. Thus, we organized the studies into clinically relevant categories and presented key aspects in tables. Due to the heterogeneity of the studies and the controversy surrounding quality scores,[5] we did not generate summary scores of study quality. Instead, we evaluated and reported key design elements that had the potential to bias the results. To recognize the more comprehensive studies in the field, we developed by consensus a set of characteristics that distinguished studies with lower risk of bias. These characteristics are shown and defined in Table 1.

General Characteristics of Included Studies
First Author and Publication Year Alarm Review Method Indicators of Potential Bias for Observational Studies Indicators of Potential Bias for Intervention Studies
Monitor System Direct Observation Medical Record Review Rhythm Annotation Video Observation Remote Monitoring Staff Medical Device Industry Involved Two Independent Reviewers At Least 1 Reviewer Is a Clinical Expert Reviewer Not Simultaneously in Patient Care Clear Definition of Alarm Actionability Census Included Statistical Testing or QI SPC Methods Fidelity Assessed Safety Assessed Lower Risk of Bias
  • NOTE: Lower risk of bias for observational studies required all of the following characteristics be reported: (1) two independent reviewers of alarms, (2) at least 1 clinical expert reviewer (ie, physician or nurse), (3) the reviewer was not simultaneously involved in clinical care, and (4) there was a clear definition of alarm actionability provided in the article. These indicators assess detection bias, observer bias, analytical bias, and reporting bias and were derived from the Meta‐analysis of Observational Studies in Epidemiology checklist.[5] Lower risk of bias for intervention studies required all of the following characteristics be reported: (1) patient census accounted for in analysis, (2) formal statistical testing of effect or statistical process control methods used to evaluate effect, (3) intervention fidelity measured and reported (defined broadly as some measurement of whether the intervention was delivered as intended), and (4) relevant patient safety outcomes measured and reported (eg, the rate of code blue events before and after the intervention was implemented). These indicators assess reporting bias and internal validity bias and were derived from the Downs and Black checklist.[42] Monitor system: alarm data were electronically collected directly from the physiologic monitors and saved on a computer device through software such as BedMasterEx. Direct observation: an in‐person observer, such as a research assistant or a nurse, takes note of the alarm data and/or responses to alarms. Medical record review: data on alarms and/or responses to alarms were extracted from the patient medical records. Rhythm annotation: data on waveforms from cardiac monitors were collected and saved on a computer device through software such as BedMasterEx. Video observation: video cameras were set up in the patient's room and recorded data on alarms and/or responses to alarms. Remote monitor staff: clinicians situated at a remote location observe the patient via video camera and may be able to communicate with the patient or the patient's assigned nurse. Abbreviations: QI, quality improvement; RN, registered nurse; SPC, statistical process control. *Monitor system + RN interrogation. Assigned nurse making observations. Monitor from central station. Alarm outcome reported using run chart, and fidelity outcomes presented using statistical process control charts.

Adult Observational
Atzema 2006[7] ✓* ✓ ✓
Billinghurst 2003[8] ✓ ✓ ✓ ✓
Biot 2000[9] ✓ ✓ ✓ ✓
Chambrin 1999[10] ✓ ✓ ✓ ✓
Drew 2014[11] ✓ ✓ ✓ ✓ ✓ ✓ ✓ ✓
Gazarian 2014[12] ✓ ✓ ✓ ✓ ✓
Grges 2009[13] ✓ ✓ ✓ ✓
Gross 2011[15] ✓ ✓ ✓ ✓ ✓ ✓ ✓
Inokuchi 2013[14] ✓ ✓ ✓ ✓ ✓ ✓ ✓
Koski 1990[16] ✓ ✓ ✓ ✓
Morales Snchez 2014[17] ✓ ✓ ✓ ✓
Pergher 2014[18] ✓ ✓
Siebig 2010[19] ✓ ✓ ✓ ✓ ✓ ✓
Voepel‐Lewis 2013[20] ✓ ✓ ✓ ✓
Way 2014[21] ✓ ✓ ✓
Pediatric Observational
Bonafide 2015[22] ✓ ✓ ✓ ✓ ✓ ✓ ✓
Lawless 1994[23] ✓ ✓
Rosman 2013[24] ✓ ✓ ✓ ✓ ✓
Talley 2011[25] ✓ ✓ ✓ ✓ ✓
Tsien 1997[26] ✓ ✓ ✓ ✓
van Pul 2015[27] ✓
Varpio 2012[28] ✓ ✓ ✓ ✓
Mixed Adult and Pediatric Observational
O'Carroll 1986[29] ✓
Wiklund 1994[30] ✓ ✓ ✓ ✓
Adult Intervention
Albert 2015[32] ✓ ✓ ✓ ✓ ✓ ✓ ✓
Cvach 2013[33] ✓ ✓
Cvach 2014[34] ✓ ✓
Graham 2010[35] ✓
Rheineck‐Leyssius 1997[36] ✓ ✓ ✓ ✓ ✓ ✓
Taenzer 2010[31] ✓ ✓ ✓ ✓ ✓ ✓
Whalen 2014[37] ✓ ✓ ✓
Pediatric Intervention
Dandoy 2014[38] ✓ ✓ ✓ ✓ ✓ ✓

For the purposes of this review, we defined nonactionable alarms as including both invalid (false) alarms that do not that accurately represent the physiologic status of the patient and alarms that are valid but do not warrant special attention or clinical intervention (nuisance alarms). We did not separate out invalid alarms due to the tremendous variation between studies in how validity was measured.

RESULTS

Study Selection

Search results produced 4629 articles (see the flow diagram in the Supporting Information in the online version of this article), of which 32 articles were eligible: 24 observational studies describing alarm characteristics and 8 studies describing interventions to reduce alarm frequency.

Observational Study Characteristics

Characteristics of included studies are shown in Table 1. Of the 24 observational studies,[7, 8, 9, 10, 11, 12, 13, 14, 15, 16, 17, 18, 19, 20, 21, 22, 23, 24, 25, 26, 27, 28, 29, 30] 15 included adult patients,[7, 8, 9, 10, 11, 12, 13, 14, 15, 16, 17, 18, 19, 20, 21] 7 included pediatric patients,[22, 23, 24, 25, 26, 27, 28] and 2 included both adult and pediatric patients.[29, 30] All were single‐hospital studies, except for 1 study by Chambrin and colleagues[10] that included 5 sites. The number of patient‐hours examined in each study ranged from 60 to 113,880.[7, 8, 9, 10, 11, 13, 14, 15, 16, 18, 19, 20, 21, 22, 23, 24, 25, 26, 27, 29, 30] Hospital settings included ICUs (n = 16),[9, 10, 11, 13, 14, 16, 17, 18, 19, 22, 23, 24, 25, 26, 27, 29] general wards (n = 5),[12, 15, 20, 22, 28] EDs (n = 2),[7, 21] postanesthesia care unit (PACU) (n = 1),[30] and cardiac care unit (CCU) (n = 1).[8] Studies varied in the type of physiologic signals recorded and data collection methods, ranging from direct observation by a nurse who was simultaneously caring for patients[29] to video recording with expert review.[14, 19, 22] Four observational studies met the criteria for lower risk of bias.[11, 14, 15, 22]

Intervention Study Characteristics

Of the 8 intervention studies, 7 included adult patients,[31, 32, 33, 34, 35, 36, 37] and 1 included pediatric patients.[38] All were single‐hospital studies; 6 were quasi‐experimental[31, 33, 34, 35, 37, 38] and 2 were experimental.[32, 36] Settings included progressive care units (n = 3),[33, 34, 35] CCUs (n = 3),[32, 33, 37] wards (n = 2),[31, 38] PACU (n = 1),[36] and a step‐down unit (n = 1).[32] All except 1 study[32] used the monitoring system to record alarm data. Several studies evaluated multicomponent interventions that included combinations of the following: widening alarm parameters,[31, 35, 36, 37, 38] instituting alarm delays,[31, 34, 36, 38] reconfiguring alarm acuity,[35, 37] use of secondary notifications,[34] daily change of electrocardiographic electrodes or use of disposable electrocardiographic wires,[32, 33, 38] universal monitoring in high‐risk populations,[31] and timely discontinuation of monitoring in low‐risk populations.[38] Four intervention studies met our prespecified lower risk of bias criteria.[31, 32, 36, 38]

Proportion of Alarms Considered Actionable

Results of the observational studies are provided in Table 2. The proportion of alarms that were actionable was 1% to 26% in adult ICU settings,[9, 10, 11, 13, 14, 16, 17, 19] 20% to 36% in adult ward settings,[12, 15, 20] 17% in a mixed adult and pediatric PACU setting,[30] 3% to 13% in pediatric ICU settings,[22, 23, 24, 25, 26] and 1% in a pediatric ward setting.[22]

Results of Included Observational Studies
Signals Included
First Author and Publication Year Setting Monitored Patient‐Hours SpO2 ECG Arrhythmia ECG Parametersa Blood Pressure Total Alarms Actionable Alarms Alarm Response Lower Risk of Bias
  • NOTE: Lower risk of bias for observational studies required all of the following characteristics be reported: (1) two independent reviewers of alarms, (2) at least 1 clinical expert reviewer (i.e. physician or nurse), (3) the reviewer was not simultaneously involved in clinical care, and (4) there was a clear definition of alarm actionability provided in the article. Abbreviations: CCU, cardiac/telemetry care unit; ECG, electrocardiogram; ED, emergency department; ICU, intensive care unit; PACU, postanesthesia care unit; SpO2, oxygen saturation; VT, ventricular tachycardia.

  • Includes respiratory rate measured via ECG leads. Actionable is defined as alarms warranting special attention or clinical intervention. Valid is defined as the alarm accurately representing the physiologic status of the patient. Directly addresses relationship between alarm exposure and response time. ∥Not provided directly; estimated from description of data collection methods.

Adult
Atzema 2006[7] ED 371 ✓ 1,762 0.20%
Billinghurst 2003[8] CCU 420 ✓ 751 Not reported; 17% were valid Nurses with higher acuity patients and smaller % of valid alarms had slower response rates
Biot 2000[9] ICU 250 ✓ ✓ ✓ ✓ 3,665 3%
Chambrin 1999[10] ICU 1,971 ✓ ✓ ✓ ✓ 3,188 26%
Drew 2014[11] ICU 48,173 ✓ ✓ ✓ ✓ 2,558,760 0.3% of 3,861 VT alarms ✓
Gazarian 2014[12] Ward 54 nurse‐hours ✓ ✓ ✓ 205 22% Response to 47% of alarms
Grges 2009[13] ICU 200 ✓ ✓ ✓ ✓ 1,214 5%
Gross 2011[15] Ward 530 ✓ ✓ ✓ ✓ 4,393 20% ✓
Inokuchi 2013[14] ICU 2,697 ✓ ✓ ✓ ✓ 11,591 6% ✓
Koski 1990[16] ICU 400 ✓ ✓ 2,322 12%
Morales Snchez 2014[17] ICU 434 sessions ✓ ✓ ✓ 215 25% Response to 93% of alarms, of which 50% were within 10 seconds
Pergher 2014[18] ICU 60 ✓ 76 Not reported 72% of alarms stopped before nurse response or had >10 minutes response time
Siebig 2010[19] ICU 982 ✓ ✓ ✓ ✓ 5,934 15%
Voepel‐Lewis 2013[20] Ward 1,616 ✓ 710 36% Response time was longer for patients in highest quartile of total alarms
Way 2014[21] ED 93 ✓ ✓ ✓ ✓ 572 Not reported; 75% were valid Nurses responded to more alarms in resuscitation room vs acute care area, but response time was longer
Pediatric
Bonafide 2015[22] Ward + ICU 210 ✓ ✓ ✓ ✓ 5,070 13% PICU, 1% ward Incremental increases in response time as number of nonactionable alarms in preceding 120 minutes increased ✓
Lawless 1994[23] ICU 928 ✓ ✓ ✓ 2,176 6%
Rosman 2013[24] ICU 8,232 ✓ ✓ ✓ ✓ 54,656 4% of rhythm alarms true critical"
Talley 2011[25] ICU 1,470∥ ✓ ✓ ✓ ✓ 2,245 3%
Tsien 1997[26] ICU 298 ✓ ✓ ✓ 2,942 8%
van Pul 2015[27] ICU 113,880∥ ✓ ✓ ✓ ✓ 222,751 Not reported Assigned nurse did not respond to 6% of alarms within 45 seconds
Varpio 2012[28] Ward 49 unit‐hours ✓ ✓ ✓ ✓ 446 Not reported 70% of all alarms and 41% of crisis alarms were not responded to within 1 minute
Both
O'Carroll 1986[29] ICU 2,258∥ ✓ 284 2%
Wiklund 1994[30] PACU 207 ✓ ✓ ✓ 1,891 17%

Relationship Between Alarm Exposure and Response Time

Whereas 9 studies addressed response time,[8, 12, 17, 18, 20, 21, 22, 27, 28] only 2 evaluated the relationship between alarm burden and nurse response time.[20, 22] Voepel‐Lewis and colleagues found that nurse responses were slower to patients with the highest quartile of alarms (57.6 seconds) compared to those with the lowest (45.4 seconds) or medium (42.3 seconds) quartiles of alarms on an adult ward (P = 0.046). They did not find an association between false alarm exposure and response time.[20] Bonafide and colleagues found incremental increases in response time as the number of nonactionable alarms in the preceding 120 minutes increased (P 0.001 in the pediatric ICU, P = 0.009 on the pediatric ward).[22]

Interventions Effective in Reducing Alarms

Results of the 8 intervention studies are provided in Table 3. Three studies evaluated single interventions;[32, 33, 36] the remainder of the studies tested interventions with multiple components such that it was impossible to separate the effect of each component. Below, we have summarized study results, arranged by component. Because only 1 study focused on pediatric patients,[38] results from pediatric and adult settings are combined.

Results of Included Intervention Studies
First Author and Publication Year Design Setting Main Intervention Components Other/ Comments Key Results Results Statistically Significant? Lower Risk of Bias
Widen Default Settings Alarm Delays Reconfigure Alarm Acuity Secondary Notification ECG Changes
  • NOTE: Lower risk of bias for intervention studies required all of the following characteristics be reported: (1) patient census accounted for in analysis, (2) formal statistical testing of effect or statistical process control methods used to evaluate effect, (3) intervention fidelity measured and reported (defined broadly as some measurement of whether the intervention was delivered as intended), and (4) relevant patient safety outcomes measured and reported (eg, the rate of code blue events before and after the intervention was implemented). Abbreviations: CCU, cardiac/telemetry care unit; ECG, electrocardiogram; ICU, intensive care unit; ITS, interrupted time series; PACU, postanesthesia care unit; PCU, progressive care unit; SpO2, oxygen saturation. *Delays were part of secondary notification system only. Delays explored retrospectively only; not part of prospective evaluation. Preimplementation count not reported.

Adult
Albert 2015[32] Experimental (cluster‐randomized) CCU ✓ Disposable vs reusable wires Disposable leads had 29% fewer no‐telemetry, leads‐fail, and leads‐off alarms and similar artifact alarms ✓ ✓
Cvach 2013[33] Quasi‐experimental (before and after) CCU and PCU ✓ Daily change of electrodes 46% fewer alarms/bed/day
Cvach 2014[34] Quasi‐experimental (ITS) PCU ✓* ✓ Slope of regression line suggests decrease of 0.75 alarms/bed/day
Graham 2010[35] Quasi‐experimental (before and after) PCU ✓ ✓ 43% fewer crisis, warning, and system warning alarms on unit
Rheineck‐Leyssius 1997[36] Experimental (RCT) PACU ✓ ✓ Alarm limit of 85% had fewer alarms/patient but higher incidence of true hypoxemia for >1 minute (6% vs 2%) ✓ ✓
Taenzer 2010[31] Quasi‐experimental (before and after with concurrent controls) Ward ✓ ✓ Universal SpO2 monitoring Rescue events decreased from 3.4 to 1.2 per 1,000 discharges; transfers to ICU decreased from 5.6 to 2.9 per 1,000 patient‐days, only 4 alarms/patient‐day ✓ ✓
Whalen 2014[37] Quasi‐experimental (before and after) CCU ✓ ✓ 89% fewer audible alarms on unit ✓
Pediatric
Dandoy 2014[38] Quasi‐experimental (ITS) Ward ✓ ✓ ✓ Timely monitor discontinuation; daily change of ECG electrodes Decrease in alarms/patient‐days from 180 to 40 ✓

Widening alarm parameter default settings was evaluated in 5 studies:[31, 35, 36, 37, 38] 1 single intervention randomized controlled trial (RCT),[36] and 4 multiple‐intervention, quasi‐experimental studies.[31, 35, 37, 38] In the RCT, using a lower SpO2 limit of 85% instead of the standard 90% resulted in 61% fewer alarms. In the 4 multiple intervention studies, 1 study reported significant reductions in alarm rates (P 0.001),[37] 1 study did not report preintervention alarm rates but reported a postintervention alarm rate of 4 alarms per patient‐day,[31] and 2 studies reported reductions in alarm rates but did not report any statistical testing.[35, 38] Of the 3 studies examining patient safety, 1 study with universal monitoring reported fewer rescue events and transfers to the ICU postimplementation,[31] 1 study reported no missed acute decompensations,[38] and 1 study (the RCT) reported significantly more true hypoxemia events (P = 0.001).[36]

Alarm delays were evaluated in 4 studies:[31, 34, 36, 38] 3 multiple‐intervention, quasi‐experimental studies[31, 34, 38] and 1 retrospective analysis of data from an RCT.[36] One study combined alarm delays with widening defaults in a universal monitoring strategy and reported a postintervention alarm rate of 4 alarms per patient.[31] Another study evaluated delays as part of a secondary notification pager system and found a negatively sloping regression line that suggested a decreasing alarm rate, but did not report statistical testing.[34] The third study reported a reduction in alarm rates but did not report statistical testing.[38] The RCT compared the impact of a hypothetical 15‐second alarm delay to that of a lower SpO2 limit reduction and reported a similar reduction in alarms.[36] Of the 4 studies examining patient safety, 1 study with universal monitoring reported improvements,[31] 2 studies reported no adverse outcomes,[35, 38] and the retrospective analysis of data from the RCT reported the theoretical adverse outcome of delayed detection of sudden, severe desaturations.[36]

Reconfiguring alarm acuity was evaluated in 2 studies, both of which were multiple‐intervention quasi‐experimental studies.[35, 37] Both showed reductions in alarm rates: 1 was significant without increasing adverse events (P 0.001),[37] and the other did not report statistical testing or safety outcomes.[35]

Secondary notification of nurses using pagers was the main intervention component of 1 study incorporating delays between the alarms and the alarm pages.[34] As mentioned above, a negatively sloping regression line was displayed, but no statistical testing or safety outcomes were reported.

Disposable electrocardiographic lead wires or daily electrode changes were evaluated in 3 studies:[32, 33, 38] 1 single intervention cluster‐randomized trial[32] and 2 quasi‐experimental studies.[33, 38] In the cluster‐randomized trial, disposable lead wires were compared to reusable lead wires, with disposable lead wires having significantly fewer technical alarms for lead signal failures (P = 0.03) but a similar number of monitoring artifact alarms (P = 0.44).[32] In a single‐intervention, quasi‐experimental study, daily electrode change showed a reduction in alarms, but no statistical testing was reported.[33] One multiple‐intervention, quasi‐experimental study incorporating daily electrode change showed fewer alarms without statistical testing.[38] Of the 2 studies examining patient safety, both reported no adverse outcomes.[32, 38]

DISCUSSION

This systematic review of physiologic monitor alarms in the hospital yielded the following main findings: (1) between 74% and 99% of physiologic monitor alarms were not actionable, (2) a significant relationship between alarm exposure and nurse response time was demonstrated in 2 small observational studies, and (3) although interventions were most often studied in combination, results from the studies with lower risk of bias suggest that widening alarm parameters, implementing alarm delays, and using disposable electrocardiographic lead wires and/or changing electrodes daily are the most promising interventions for reducing alarms. Only 5 of 8 intervention studies measured intervention safety and found that widening alarm parameters and implementing alarm delays had mixed safety outcomes, whereas disposable electrocardiographic lead wires and daily electrode changes had no adverse safety outcomes.[29, 30, 34, 35, 36] Safety measures are crucial to ensuring the highest level of patient safety is met; interventions are rendered useless without ensuring actionable alarms are not disabled. The variation in results across studies likely reflects the wide range of care settings as well as differences in design and quality.

This field is still in its infancy, with 18 of the 32 articles published in the past 5 years. We anticipate improvements in quality and rigor as the field matures, as well as clinically tested interventions that incorporate smart alarms. Smart alarms integrate data from multiple physiologic signals and the patient's history to better detect physiologic changes in the patient and improve the positive predictive value of alarms. Academicindustry partnerships will be required to implement and rigorously test smart alarms and other emerging technologies in the hospital.

To our knowledge, this is the first systematic review focused on monitor alarms with specific review questions relevant to alarm fatigue. Cvach recently published an integrative review of alarm fatigue using research published through 2011.[39] Our review builds upon her work by contributing a more extensive and systematic search strategy with databases spanning nursing, medicine, and engineering, including additional languages, and including newer studies published through April 2015. In addition, we included multiple cross‐team checks in our eligibility review to ensure high sensitivity and specificity of the resulting set of studies.

Although we focused on interventions aiming to reduce alarms, there has also been important recent work focused on reducing telemetry utilization in adult hospital populations as well as work focused on reducing pulse oximetry utilization in children admitted with respiratory conditions. Dressler and colleagues reported an immediate and sustained reduction in telemetry utilization in hospitalized adults upon redesign of cardiac telemetry order sets to include the clinical indication, which defaulted to the American Heart Association guideline‐recommended telemetry duration.[40] Instructions for bedside nurses were also included in the order set to facilitate appropriate telemetry discontinuation. Schondelmeyer and colleagues reported reductions in continuous pulse oximetry utilization in hospitalized children with asthma and bronchiolitis upon introduction of a multifaceted quality improvement program that included provider education, a nurse handoff checklist, and discontinuation criteria incorporated into order sets.[41]

Limitations of This Review and the Underlying Body of Work

There are limitations to this systematic review and its underlying body of work. With respect to our approach to this systematic review, we focused only on monitor alarms. Numerous other medical devices generate alarms in the patient‐care environment that also can contribute to alarm fatigue and deserve equally rigorous evaluation. With respect to the underlying body of work, the quality of individual studies was generally low. For example, determinations of alarm actionability were often made by a single rater without evaluation of the reliability or validity of these determinations, and statistical testing was often missing. There were also limitations specific to intervention studies, including evaluation of nongeneralizable patient populations, failure to measure the fidelity of the interventions, inadequate measures of intervention safety, and failure to statistically evaluate alarm reductions. Finally, though not necessarily a limitation, several studies were conducted by authors involved in or funded by the medical device industry.[11, 15, 19, 31, 32] This has the potential to introduce bias, although we have no indication that the quality of the science was adversely impacted.

Moving forward, the research agenda for physiologic monitor alarms should include the following: (1) more intensive focus on evaluating the relationship between alarm exposure and response time with analysis of important mediating factors that may promote or prevent alarm fatigue, (2) emphasis on studying interventions aimed at improving alarm management using rigorous designs such as cluster‐randomized trials and trials randomized by individual participant, (3) monitoring and reporting clinically meaningful balancing measures that represent unintended consequences of disabling or delaying potentially important alarms and possibly reducing the clinicians' ability to detect true patient deterioration and intervene in a timely manner, and (4) support for transparent academicindustry partnerships to evaluate new alarm technology in real‐world settings. As evidence‐based interventions emerge, there will be new opportunities to study different implementation strategies of these interventions to optimize effectiveness.

CONCLUSIONS

The body of literature relevant to physiologic monitor alarm characteristics and alarm fatigue is limited but growing rapidly. Although we know that most alarms are not actionable and that there appears to be a relationship between alarm exposure and response time that could be caused by alarm fatigue, we cannot yet say with certainty that we know which interventions are most effective in safely reducing unnecessary alarms. Interventions that appear most promising and should be prioritized for intensive evaluation include widening alarm parameters, implementing alarm delays, and using disposable electrocardiographic lead wires and changing electrodes daily. Careful evaluation of these interventions must include systematically examining adverse patient safety consequences.

Acknowledgements

The authors thank Amogh Karnik and Micheal Sellars for their technical assistance during the review and extraction process.

Disclosures: Ms. Zander is supported by the Society of Hospital Medicine Student Hospitalist Scholar Grant. Dr. Bonafide and Ms. Stemler are supported by the National Heart, Lung, and Blood Institute of the National Institutes of Health under award number K23HL116427. The content is solely the responsibility of the authors and does not necessarily represent the official views of the National Institutes of Health. The authors report no conflicts of interest.

References
  1. National Patient Safety Goals Effective January 1, 2015. The Joint Commission Web site. http://www.jointcommission.org/assets/1/6/2015_NPSG_HAP.pdf. Accessed July 17, 2015.
  2. ECRI Institute. 2015 Top 10 Health Technology Hazards. Available at: https://www.ecri.org/Pages/2015‐Hazards.aspx. Accessed June 23, 2015.
  3. Sendelbach S, Funk M. Alarm fatigue: a patient safety concern. AACN Adv Crit Care. 2013;24(4):378–386.
  4. Chopra V, McMahon LF. Redesigning hospital alarms for patient safety: alarmed and potentially dangerous. JAMA. 2014;311(12):1199–1200.
  5. Stroup DF, Berlin JA, Morton SC, et al. Meta‐analysis of observational studies in epidemiology: a proposal for reporting. Meta‐analysis Of Observational Studies in Epidemiology (MOOSE) Group. JAMA. 2000;283(15):2008–2012.
  6. Moher D, Liberati A, Tetzlaff J, Altman DG; PRISMA Group. Preferred reporting items for systematic reviews and meta‐analyses: the PRISMA statement. Ann Intern Med. 2009;151(4):264–269, W64.
  7. Atzema C, Schull MJ, Borgundvaag B, Slaughter GRD, Lee CK. ALARMED: adverse events in low‐risk patients with chest pain receiving continuous electrocardiographic monitoring in the emergency department. A pilot study. Am J Emerg Med. 2006;24:62–67.
  8. Billinghurst F, Morgan B, Arthur HM. Patient and nurse‐related implications of remote cardiac telemetry. Clin Nurs Res. 2003;12(4):356–370.
  9. Biot L, Carry PY, Perdrix JP, Eberhard A, Baconnier P. Clinical evaluation of alarm efficiency in intensive care [in French]. Ann Fr Anesth Reanim. 2000;19:459–466.
  10. Chambrin MC, Ravaux P, Calvelo‐Aros D, Jaborska A, Chopin C, Boniface B. Multicentric study of monitoring alarms in the adult intensive care unit (ICU): a descriptive analysis. Intensive Care Med. 1999;25:1360–1366.
  11. Drew BJ, Harris P, Zègre‐Hemsey JK, et al. Insights into the problem of alarm fatigue with physiologic monitor devices: a comprehensive observational study of consecutive intensive care unit patients. PloS One. 2014;9(10):e110274.
  12. Gazarian PK. Nurses' response to frequency and types of electrocardiography alarms in a non‐ critical care setting: a descriptive study. Int J Nurs Stud. 2014;51(2):190–197.
  13. Görges M, Markewitz BA, Westenskow DR. Improving alarm performance in the medical intensive care unit using delays and clinical context. Anesth Analg. 2009;108:1546–1552.
  14. Inokuchi R, Sato H, Nanjo Y, et al. The proportion of clinically relevant alarms decreases as patient clinical severity decreases in intensive care units: a pilot study. BMJ Open. 2013;3(9):e003354–e003354.
  15. Gross B, Dahl D, Nielsen L. Physiologic monitoring alarm load on medical/surgical floors of a community hospital. Biomed Instrum Technol. 2011;45:29–36.
  16. Koski EM, Mäkivirta A, Sukuvaara T, Kari A. Frequency and reliability of alarms in the monitoring of cardiac postoperative patients. Int J Clin Monit Comput. 1990;7(2):129–133.
  17. Morales Sánchez C, Murillo Pérez MA, Torrente Vela S, et al. Audit of the bedside monitor alarms in a critical care unit [in Spanish]. Enferm Intensiva. 2014;25(3):83–90.
  18. Pergher AK, Silva RCL. Stimulus‐response time to invasive blood pressure alarms: implications for the safety of critical‐care patients. Rev Gaúcha Enferm. 2014;35(2):135–141.
  19. Siebig S, Kuhls S, Imhoff M, Gather U, Scholmerich J, Wrede CE. Intensive care unit alarms— how many do we need? Crit Care Med. 2010;38:451–456.
  20. Voepel‐Lewis T, Parker ML, Burke CN, et al. Pulse oximetry desaturation alarms on a general postoperative adult unit: a prospective observational study of nurse response time. Int J Nurs Stud. 2013;50(10):1351–1358.
  21. Way RB, Beer SA, Wilson SJ. Whats that noise? Bedside monitoring in the Emergency Department. Int Emerg Nurs. 2014;22(4):197–201.
  22. Bonafide CP, Lin R, Zander M, et al. Association between exposure to nonactionable physiologic monitor alarms and response time in a children's hospital. J Hosp Med. 2015;10(6):345–351.
  23. Lawless ST. Crying wolf: false alarms in a pediatric intensive care unit. Crit Care Med. 1994;22(6):981–985.
  24. Rosman EC, Blaufox AD, Menco A, Trope R, Seiden HS. What are we missing? Arrhythmia detection in the pediatric intensive care unit. J Pediatr. 2013;163(2):511–514.
  25. Talley LB, Hooper J, Jacobs B, et al. Cardiopulmonary monitors and clinically significant events in critically ill children. Biomed Instrum Technol. 2011;45(s1):38–45.
  26. Tsien CL, Fackler JC. Poor prognosis for existing monitors in the intensive care unit. Crit Care Med. 1997;25:614–619.
  27. Pul C, Mortel H, Bogaart J, Mohns T, Andriessen P. Safe patient monitoring is challenging but still feasible in a neonatal intensive care unit with single family rooms. Acta Paediatr Oslo Nor 1992. 2015;104(6):e247–e254.
  28. Varpio L, Kuziemsky C, Macdonald C, King WJ. The helpful or hindering effects of in‐hospital patient monitor alarms on nurses: a qualitative analysis. CIN Comput Inform Nurs. 2012;30(4):210–217.
  29. O'Carroll T. Survey of alarms in an intensive therapy unit. Anaesthesia. 1986;41(7):742–744.
  30. Wiklund L, Hök B, Ståhl K, Jordeby‐Jönsson A. Postanesthesia monitoring revisited: frequency of true and false alarms from different monitoring devices. J Clin Anesth. 1994;6(3):182–188.
  31. Taenzer AH, Pyke JB, McGrath SP, Blike GT. Impact of pulse oximetry surveillance on rescue events and intensive care unit transfers: a before‐and‐after concurrence study. Anesthesiology. 2010;112(2):282–287.
  32. Albert NM, Murray T, Bena JF, et al. Differences in alarm events between disposable and reusable electrocardiography lead wires. Am J Crit Care. 2015;24(1):67–74.
  33. Cvach MM, Biggs M, Rothwell KJ, Charles‐Hudson C. Daily electrode change and effect on cardiac monitor alarms: an evidence‐based practice approach. J Nurs Care Qual. 2013;28:265–271.
  34. Cvach MM, Frank RJ, Doyle P, Stevens ZK. Use of pagers with an alarm escalation system to reduce cardiac monitor alarm signals. J Nurs Care Qual. 2014;29(1):9–18.
  35. Graham KC, Cvach M. Monitor alarm fatigue: standardizing use of physiological monitors and decreasing nuisance alarms. Am J Crit Care. 2010;19:28–34.
  36. Rheineck‐Leyssius AT, Kalkman CJ. Influence of pulse oximeter lower alarm limit on the incidence of hypoxaemia in the recovery room. Br J Anaesth. 1997;79(4):460–464.
  37. Whalen DA, Covelle PM, Piepenbrink JC, Villanova KL, Cuneo CL, Awtry EH. Novel approach to cardiac alarm management on telemetry units. J Cardiovasc Nurs. 2014;29(5):E13–E22.
  38. Dandoy CE, Davies SM, Flesch L, et al. A team‐based approach to reducing cardiac monitor alarms. Pediatrics. 2014;134(6):e1686–e1694.
  39. Cvach M. Monitor alarm fatigue: an integrative review. Biomed Instrum Technol. 2012;46(4):268–277.
  40. Dressler R, Dryer MM, Coletti C, Mahoney D, Doorey AJ. Altering overuse of cardiac telemetry in non‐intensive care unit settings by hardwiring the use of American Heart Association guidelines. JAMA Intern Med. 2014;174(11):1852–1854.
  41. Schondelmeyer AC, Simmons JM, Statile AM, et al. Using quality improvement to reduce continuous pulse oximetry use in children with wheezing. Pediatrics. 2015;135(4):e1044–e1051.
  42. Downs SH, Black N. The feasibility of creating a checklist for the assessment of the methodological quality both of randomised and non‐randomised studies of health care interventions. J Epidemiol Community Health. 1998;52(6):377–384.
References
  1. National Patient Safety Goals Effective January 1, 2015. The Joint Commission Web site. http://www.jointcommission.org/assets/1/6/2015_NPSG_HAP.pdf. Accessed July 17, 2015.
  2. ECRI Institute. 2015 Top 10 Health Technology Hazards. Available at: https://www.ecri.org/Pages/2015‐Hazards.aspx. Accessed June 23, 2015.
  3. Sendelbach S, Funk M. Alarm fatigue: a patient safety concern. AACN Adv Crit Care. 2013;24(4):378–386.
  4. Chopra V, McMahon LF. Redesigning hospital alarms for patient safety: alarmed and potentially dangerous. JAMA. 2014;311(12):1199–1200.
  5. Stroup DF, Berlin JA, Morton SC, et al. Meta‐analysis of observational studies in epidemiology: a proposal for reporting. Meta‐analysis Of Observational Studies in Epidemiology (MOOSE) Group. JAMA. 2000;283(15):2008–2012.
  6. Moher D, Liberati A, Tetzlaff J, Altman DG; PRISMA Group. Preferred reporting items for systematic reviews and meta‐analyses: the PRISMA statement. Ann Intern Med. 2009;151(4):264–269, W64.
  7. Atzema C, Schull MJ, Borgundvaag B, Slaughter GRD, Lee CK. ALARMED: adverse events in low‐risk patients with chest pain receiving continuous electrocardiographic monitoring in the emergency department. A pilot study. Am J Emerg Med. 2006;24:62–67.
  8. Billinghurst F, Morgan B, Arthur HM. Patient and nurse‐related implications of remote cardiac telemetry. Clin Nurs Res. 2003;12(4):356–370.
  9. Biot L, Carry PY, Perdrix JP, Eberhard A, Baconnier P. Clinical evaluation of alarm efficiency in intensive care [in French]. Ann Fr Anesth Reanim. 2000;19:459–466.
  10. Chambrin MC, Ravaux P, Calvelo‐Aros D, Jaborska A, Chopin C, Boniface B. Multicentric study of monitoring alarms in the adult intensive care unit (ICU): a descriptive analysis. Intensive Care Med. 1999;25:1360–1366.
  11. Drew BJ, Harris P, Zègre‐Hemsey JK, et al. Insights into the problem of alarm fatigue with physiologic monitor devices: a comprehensive observational study of consecutive intensive care unit patients. PloS One. 2014;9(10):e110274.
  12. Gazarian PK. Nurses' response to frequency and types of electrocardiography alarms in a non‐ critical care setting: a descriptive study. Int J Nurs Stud. 2014;51(2):190–197.
  13. Görges M, Markewitz BA, Westenskow DR. Improving alarm performance in the medical intensive care unit using delays and clinical context. Anesth Analg. 2009;108:1546–1552.
  14. Inokuchi R, Sato H, Nanjo Y, et al. The proportion of clinically relevant alarms decreases as patient clinical severity decreases in intensive care units: a pilot study. BMJ Open. 2013;3(9):e003354–e003354.
  15. Gross B, Dahl D, Nielsen L. Physiologic monitoring alarm load on medical/surgical floors of a community hospital. Biomed Instrum Technol. 2011;45:29–36.
  16. Koski EM, Mäkivirta A, Sukuvaara T, Kari A. Frequency and reliability of alarms in the monitoring of cardiac postoperative patients. Int J Clin Monit Comput. 1990;7(2):129–133.
  17. Morales Sánchez C, Murillo Pérez MA, Torrente Vela S, et al. Audit of the bedside monitor alarms in a critical care unit [in Spanish]. Enferm Intensiva. 2014;25(3):83–90.
  18. Pergher AK, Silva RCL. Stimulus‐response time to invasive blood pressure alarms: implications for the safety of critical‐care patients. Rev Gaúcha Enferm. 2014;35(2):135–141.
  19. Siebig S, Kuhls S, Imhoff M, Gather U, Scholmerich J, Wrede CE. Intensive care unit alarms— how many do we need? Crit Care Med. 2010;38:451–456.
  20. Voepel‐Lewis T, Parker ML, Burke CN, et al. Pulse oximetry desaturation alarms on a general postoperative adult unit: a prospective observational study of nurse response time. Int J Nurs Stud. 2013;50(10):1351–1358.
  21. Way RB, Beer SA, Wilson SJ. Whats that noise? Bedside monitoring in the Emergency Department. Int Emerg Nurs. 2014;22(4):197–201.
  22. Bonafide CP, Lin R, Zander M, et al. Association between exposure to nonactionable physiologic monitor alarms and response time in a children's hospital. J Hosp Med. 2015;10(6):345–351.
  23. Lawless ST. Crying wolf: false alarms in a pediatric intensive care unit. Crit Care Med. 1994;22(6):981–985.
  24. Rosman EC, Blaufox AD, Menco A, Trope R, Seiden HS. What are we missing? Arrhythmia detection in the pediatric intensive care unit. J Pediatr. 2013;163(2):511–514.
  25. Talley LB, Hooper J, Jacobs B, et al. Cardiopulmonary monitors and clinically significant events in critically ill children. Biomed Instrum Technol. 2011;45(s1):38–45.
  26. Tsien CL, Fackler JC. Poor prognosis for existing monitors in the intensive care unit. Crit Care Med. 1997;25:614–619.
  27. Pul C, Mortel H, Bogaart J, Mohns T, Andriessen P. Safe patient monitoring is challenging but still feasible in a neonatal intensive care unit with single family rooms. Acta Paediatr Oslo Nor 1992. 2015;104(6):e247–e254.
  28. Varpio L, Kuziemsky C, Macdonald C, King WJ. The helpful or hindering effects of in‐hospital patient monitor alarms on nurses: a qualitative analysis. CIN Comput Inform Nurs. 2012;30(4):210–217.
  29. O'Carroll T. Survey of alarms in an intensive therapy unit. Anaesthesia. 1986;41(7):742–744.
  30. Wiklund L, Hök B, Ståhl K, Jordeby‐Jönsson A. Postanesthesia monitoring revisited: frequency of true and false alarms from different monitoring devices. J Clin Anesth. 1994;6(3):182–188.
  31. Taenzer AH, Pyke JB, McGrath SP, Blike GT. Impact of pulse oximetry surveillance on rescue events and intensive care unit transfers: a before‐and‐after concurrence study. Anesthesiology. 2010;112(2):282–287.
  32. Albert NM, Murray T, Bena JF, et al. Differences in alarm events between disposable and reusable electrocardiography lead wires. Am J Crit Care. 2015;24(1):67–74.
  33. Cvach MM, Biggs M, Rothwell KJ, Charles‐Hudson C. Daily electrode change and effect on cardiac monitor alarms: an evidence‐based practice approach. J Nurs Care Qual. 2013;28:265–271.
  34. Cvach MM, Frank RJ, Doyle P, Stevens ZK. Use of pagers with an alarm escalation system to reduce cardiac monitor alarm signals. J Nurs Care Qual. 2014;29(1):9–18.
  35. Graham KC, Cvach M. Monitor alarm fatigue: standardizing use of physiological monitors and decreasing nuisance alarms. Am J Crit Care. 2010;19:28–34.
  36. Rheineck‐Leyssius AT, Kalkman CJ. Influence of pulse oximeter lower alarm limit on the incidence of hypoxaemia in the recovery room. Br J Anaesth. 1997;79(4):460–464.
  37. Whalen DA, Covelle PM, Piepenbrink JC, Villanova KL, Cuneo CL, Awtry EH. Novel approach to cardiac alarm management on telemetry units. J Cardiovasc Nurs. 2014;29(5):E13–E22.
  38. Dandoy CE, Davies SM, Flesch L, et al. A team‐based approach to reducing cardiac monitor alarms. Pediatrics. 2014;134(6):e1686–e1694.
  39. Cvach M. Monitor alarm fatigue: an integrative review. Biomed Instrum Technol. 2012;46(4):268–277.
  40. Dressler R, Dryer MM, Coletti C, Mahoney D, Doorey AJ. Altering overuse of cardiac telemetry in non‐intensive care unit settings by hardwiring the use of American Heart Association guidelines. JAMA Intern Med. 2014;174(11):1852–1854.
  41. Schondelmeyer AC, Simmons JM, Statile AM, et al. Using quality improvement to reduce continuous pulse oximetry use in children with wheezing. Pediatrics. 2015;135(4):e1044–e1051.
  42. Downs SH, Black N. The feasibility of creating a checklist for the assessment of the methodological quality both of randomised and non‐randomised studies of health care interventions. J Epidemiol Community Health. 1998;52(6):377–384.
Issue
Journal of Hospital Medicine - 11(2)
Issue
Journal of Hospital Medicine - 11(2)
Page Number
136-144
Page Number
136-144
Article Type
Display Headline
Systematic Review of Physiologic Monitor Alarm Characteristics and Pragmatic Interventions to Reduce Alarm Frequency
Display Headline
Systematic Review of Physiologic Monitor Alarm Characteristics and Pragmatic Interventions to Reduce Alarm Frequency
Sections
Article Source
© 2015 Society of Hospital Medicine
Disallow All Ads
Correspondence Location
Address for correspondence and reprint requests: Christopher P. Bonafide, MD, MSCE, The Children's Hospital of Philadelphia, 3401 Civic Center Blvd., Philadelphia, PA 19104; Telephone: 267‐426‐2901; E‐mail: [email protected]
Content Gating
Gated (full article locked unless allowed per User)
Gating Strategy
First Peek Free
Article PDF Media
Media Files